Interference from Deuterated IS [Bioanalytics]
Dear VKB2207,
Sure. It's called the intercept
Certainly not. It is taken into consideration when calculating the calibration curve parameters, unless you are forcing the intercept through 0 (which is not to be done in bioanalysis).
Not necessarily at high level of concentrations. Could indeed be problematic closer to the LLOQ.
❝ You can probably evaluate the % contribution from IS at RT of drug
Sure. It's called the intercept

❝ and then apply this factor in CC, QC nominal concentrations.
Certainly not. It is taken into consideration when calculating the calibration curve parameters, unless you are forcing the intercept through 0 (which is not to be done in bioanalysis).
❝ Hence measured concentrations would be inaccurate.
Not necessarily at high level of concentrations. Could indeed be problematic closer to the LLOQ.
—
Regards
Ohlbe
Regards
Ohlbe