Interference from Deuterated IS [Bioanalytics]

posted by VKB2207 – Portugal, 2019-01-10 20:48 (829 d 03:40 ago) – Posting: # 19767
Views: 2,545

Hi,

Suggest you not to go with MV in this scenario as IS is contributing and you have no exact idea how much it is contributing. Hence measured concentrations would be inaccurate.

You can probably evaluate the % contribution from IS at RT of drug and then apply this factor in CC, QC nominal concentrations.

Hope it helps.