Selectivity [Bioanalytics]

posted by Ohlbe – France, 2018-11-13 13:56 (1045 d 19:21 ago) – Posting: # 19596
Views: 2,238

Dear Avinash,

» Which means? Should we run Selectivity samples with calibrators?

Not necessarily a full calibration curve, but at least samples spiked at the LLOQ, so that you can compare the signal in the blanks to the signal at the LLOQ.

Regards
Ohlbe

Complete thread: