Yes it went a little fast [Bioanalytics]

posted by nobody – 2018-11-07 19:30 (2140 d 08:14 ago) – Posting: # 19544
Views: 8,000

Formally, the limit of detection (LOD) is defined by the S:N ratio. The peak you can definitely distinguish from noise (Limit of blank) with your method. In bioanalytics the LLOQ is by definition the concentration you can determine with P/A of 20%/+/-20% (by a convention method, i.e spiked samples with n=......).

If you (or anybody else in the room...) think your LOD is higher than the LLOQ, you have a problem, in deed. Otherwise... see Helllcourages statement.

Kindest regards, nobody