Dear ElMaestro,

❝ is there any method published anywhere which can be realistically thought to have no matrix effect?

❝ Not just absence of mentioned effect, but truly verified absence of one and how would one go about proving it?


Define "no matrix effect". Same as concentration = 0: truly 0, or BLQ ? No matrix effect, or negligible ? Considering the analytical variability, I would consider it strictly impossible to demonstrate 0.000 matrix effect.

❝ If you had a gun pointed to your head and someone asks you to develop an LC-MS/MS method without matrix effect for, say, Duloxetine or Simvastatin where would you start?


Our Father who art in heaven...

I would start with:

  1. using matrix obtained the same way as subject samples (vacutainers, not commercial plasma collected in blood collection bags full of plasticisers)
  2. a real, proper sample cleaning (SPE or LLE, with a real development to really clean up, without too much loss in recovery. For instance, MTBE usually gives you a very good recovery of your drug... but also of quite a lot of things you'd like to get rid of)
  3. a real chromatography (not a 90 seconds run on a 30 mm column)
  4. comparing APCI and ESI. APCI is supposed to be less sensitive to matrix effects. Unfortunately, it is also less sensitive, full stop. May be problematic if a very low LLOQ is required (would not work for a number of the molecules you're working on).

And that's if I limit the definition of matrix effects to differences in detector response. You could expand the definition to include differences in recovery.

Regards
Ohlbe

Complete thread:

UA Flag
Activity
 Admin contact
22,957 posts in 4,819 threads, 1,636 registered users;
108 visitors (0 registered, 108 guests [including 6 identified bots]).
Forum time: 16:05 CET (Europe/Vienna)

With four parameters I can fit an elephant,
and with five I can make him wiggle his trunk.    John von Neumann

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5