recoveries more than 100% [Bioanalytics]

posted by Ohlbe – France, 2018-09-03 19:21 (1180 d 06:30 ago) – Posting: # 19231
Views: 3,805

Dear Deepak,

» The % recovery was determined by comparing the mean peak area of 6 replicates of extracted quality control samples at High, Middle and Low concentrations against respective mean peak area of 6 replicates of un-extracted quality control samples at High, Middle and Low concentrations.

Meaning that what you observe is a mix of extraction recovery and matrix effect. If you want to see the true extraction recovery you have to compare to samples spiked post-extraction. You can combine with matrix effects experiments.

» Internal standard used was deuterated.

Makes no difference. You are looking at peak areas, not peak area ratios.

» Kindly confirm recoveries markedly higher than 100% for LQC, MQC2 and MQC1 are acceptable as per USFDA guidelines??

Did you see any limit for absolute recovery in either direction in the FDA guidance ?

» Also specify reason for higher recoveries.

Multiple. The most probable: analytical variability (your results are well within the 15 % acceptance limits for precision and accuracy). Other possibilities include ion enhancement, sample preparation, calculation errors...

Regards
Ohlbe

Complete thread:

Activity
 Admin contact
21,778 posts in 4,555 threads, 1,547 registered users;
online 5 (0 registered, 5 guests [including 5 identified bots]).
Forum time: Saturday 00:51 CET (Europe/Vienna)

Every new technology to me is like a newborn baby –
and you think that it’s become president
or it’s cure cancer or win a nobel prize.
But in the end, you’re perfectly happy
when it just stays out of jail […].    Eric Betzig

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5