Freshly spiked calibration standard in validation [Bioanalytics]

posted by Obinoscopy  – USA, 2018-05-02 12:45 (1568 d 06:10 ago) – Posting: # 18750
Views: 1,433

Hi,

» So shall we use stored calibration standards for evaluation of PA during validation, if we prove Long term matrix stability at the end of validation?

Yes you can, as long as you demonstrated the long term matrix stability of your analyte.

But the critical question is this: which CC did you use in demonstrating long term stability? Was the curve gotten from freshly prepared calibration standards? This is because only CCs from freshly prepared standards are to be used for stability determination.

Regards,

Scopy

Complete thread:

UA Flag
Activity
 Admin contact
22,305 posts in 4,668 threads, 1,587 registered users;
online 14 (1 registered, 13 guests [including 10 identified bots]).
Forum time: Wednesday 18:56 CEST (Europe/Vienna)

There is no point in being precise when you don’t know
what you’re talking about.    attributed to John Tukey

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5