Serial dilutions? Don’t! [Study As­sess­ment]

posted by Helmut Homepage – Vienna, Austria, 2017-10-30 18:53 (2362 d 09:12 ago) – Posting: # 17955
Views: 7,899

Hi öberster größter Meister,

❝ dilute the sample you already diluted; use the already validated dilution factor and inject. Continue doing that until you get a measurable concentration.


Really? Quoting the BMV-GL:

4.1.7. Dilution integrity
Dilution of samples should not affect the accuracy and precision. If applicable, dilution integrity should be demonstrated by spiking the matrix with an analyte concentration above the ULOQ and diluting this sample with blank matrix (at least five determinations per dilution factor). Accuracy and precision should be within the set criteria, i.e. within ±15%. Dilution integrity should cover the dilution applied to the study samples.

(my emphasis)

If you want to dilute ×2, ×4, ×8 those are the dilution factors – not serial ones like 1×2, 2x2, 2x2x2. Imagine a propagation of error which results in an accuracy of 90% in each of the steps. If done properly (only one dilution), you pass the requirements (within ±15%). With serial dilutions you would fail already in the second step (81% outside ±15%) and even more in the third (73%). But: Since you didn’t validate it, you falsely assume that the accuracy is still 90%. You don’t want to go there.
Of course, you could validate serial dilutions as well but I don’t see any rationale for doing so.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
22,987 posts in 4,824 threads, 1,669 registered users;
87 visitors (0 registered, 87 guests [including 5 identified bots]).
Forum time: 05:05 CEST (Europe/Vienna)

The only way to comprehend what mathematicians mean by Infinity
is to contemplate the extent of human stupidity.    Voltaire

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5