TSDs for RSABE/ABEL [Two-Stage / GS Designs]
Since Two-Stage Designs were mentioned in various guidelines (EMA, FDA,…) sponsors want to combine them with options for HVDs/HVDPs (EMA: ABEL, FDA: RSABE). I do understand that this a tempting idea. What I have heard at recent conferences gives me the impression that sponsors already performed such studies. I don’t know whether any one of them was accepted. I know of one leading to a deficiency letter.
I will summarize the caveats.
- Reference-scaling might lead to an inflation of the type I error (TIE) – caused by modifying the Null-Hypothesis “in the face of the data”. The method is intrinsically sequential:
- Estimate the intra-subject variability of the reference.
- Eventually widen the acceptance-range (EMA) or assess the linearized criterion (FDA).
[…] the equivalence limits for the scaled ABE involve the unknown intra-subject variability. As a result, the equivalence limits become random variables and are not fixed constants and the variability introduced by scaling will have an impact on the type I error rate. Therefore, the attempt to use the scaled ABE for resolution of high variable drug products will face the similar issues and challenges that the individual and population bioequivalence encountered in the 1990s […]. These issues must be satisfactorily addressed before the scaled ABE can be implemented into regulatory framework.
Though attempts have been made to adjust α based on the first stage’s data5,6 regulatory acceptance is problematic, since – at least for the EMA – the adjusted level has to be pre-specified in the protocol.
If one wants to pre-specify an adjusted α (GL…) one has to (empirically) find one which keeps the TIE for all possible combinations of stage 1 sample sizes and CVs. Currently I use a matrix with ~1,000 combos (n1 and CV with a step size of 2). For every grid-point one has to simulate 106 BE-studies for the TIE (slow convergence) and 105 for power. This requires 1.1×109 simulations.
Contrary to 2×2×2 and parallel designs – where power can be directly calculated – for scaling we need simulations since the limits are not fixed, we have a restriction on the GMR, and an upper cap at 50% for EMA. The convergence is fine, but we still need 105 simulations within every simulated study. This leads to ~1014 (110,000,000,000,000‼) simulations overall.
To give you an idea:
run0 <- 1e7; run <- 1:run0
ptm <- proc.time() # time the loop's overhead
for(j in seq_along(run)) { }
t0 <- as.numeric(1000*(proc.time()[3]-ptm[3])/run0)
run1 <- 1e4; run <- 1:run1
ptm <- proc.time()
for(j in seq_along(run)) {
power.TOST(CV=0.4, theta0=0.9, design="2x2x2", n=134, method="nct") }
t1 <- as.numeric(1000*(proc.time()[3]-ptm[3])/run1-t0)
run2 <- 1e3; run <- 1:run2
ptm <- proc.time() # ABEL (EMA)
for(j in seq_along(run)) {
power.scABEL(CV=0.4, theta0=0.9, design="2x2x4", n=30, nsims=1e5) }
t2 <- as.numeric(1000*(proc.time()[3]-ptm[3])/run2-t0)
run3 <- 1e3; run <- 1:run3
ptm <- proc.time() # RSABE (FDA)
for(j in seq_along(run)) {
power.RSABE(CV=0.4, theta0=0.9, design="2x2x4", n=24, nsims=1e5) }
t3 <- as.numeric(1000*(proc.time()[3]-ptm[3])/run3-t0)
cat("Runtimes of PowerTOST's functions:\n",
sprintf("%s %5.2f %s", "2x2x2 ABE :", t1, "ms\n"),
sprintf("%s %5.2f %s (%1.0fx slower)%s", "2x2x4 ABEL (EMA):", t2, "ms", t2/t1, "\n"),
sprintf("%s %5.2f %s (%1.0fx slower)%s", "2x2x4 RSABE (FDA):", t3, "ms", t3/t1, "\n"))
Gives on my machine:
Runtimes of PowerTOST's functions:
2x2x2 ABE : 2.22 ms
2x2x4 ABEL (EMA): 82.82 ms (37x slower)
2x2x4 RSABE (FDA): 80.45 ms (36x slower)
Detlew did a great job. The scaled power-functions are not by a factor of 105 slower – only ~40times. If you have a lot of time, go ahead and become famous.
Furthermore, the intra-subject variabilities of test and reference must not be identical. Keep that in mind.
Edit: Easier with the package microbenchmark:
library(microbenchmark)
library(PowerTOST)
microbenchmark(
power.TOST(CV=0.4, theta0=0.9, design="2x2x2", n=134, method="nct"),
power.scABEL(CV=0.4, theta0=0.9, design="2x2x4", n= 30, nsims=1e5),
power.RSABE(CV=0.4, theta0=0.9, design="2x2x4", n= 24, nsims=1e5))
On my machine (slow because I have 8 R-sessions running; CPU-load >95%):
Unit: milliseconds
expr min lq mean median uq max neval
power.TOST() 3.703643 4.336665 5.82738 4.784337 5.475469 67.81569 100
power.scABEL() 84.496413 91.930565 120.51301 97.457502 156.742944 176.92144 100
power.RSABE() 91.391727 103.207520 139.51786 155.408528 167.997411 205.35759 100
- References:
- Chow S-C, Liu J-p. Design and Analysis of Bioavailability and Bioequivalence Studies. Boca Raton: Chapman & Hall/CRC; 3rd ed 2009: p. 598.
- Endrényi L, Tóthfalusi L. Regulatory conditions for the determination of bioequivalence of highly variable drugs. J Pharm Pharm Sci. 2009;12:138–49. free resource.
- Wonnemann M, Frömke C, Koch A. Inflation of the Type I Error: Investigations on Regulatory Recommendations for Bioequivalence of Highly Variable Drugs. Pharm Res. 2015;32(1):135–43. doi:10.1007/s11095-014-1450-z.
- Schütz H. Two-stage designs in bioequivalence trials. Eur J Clin Pharmacol. 2015;71(3):271-81. doi:10.1007/s00228-015-1806-2.
- Fuglsang A. Controlling type I errors for two-stage bioequivalence study designs. Clin Res Reg Aff. 2011;28(4):100–5. doi:10.3109/10601333.2011.631547.
- Kieser M, Rauch G. Two-stage designs for cross-over bioequivalence trials. Stat Med. 2015;34(16):2403–16. doi:10.1002/sim.6487.
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
Helmut Schütz
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- TSDs for RSABE/ABELHelmut 2015-05-06 02:17
- TSDs for RSABE/ABEL nobody 2015-05-06 11:13
- TSDs for RSABE/ABEL Helmut 2015-05-06 11:55
- TSDs for RSABE/ABEL nobody 2015-05-06 12:26
- TSDs for RSABE/ABEL Helmut 2015-05-06 11:55
- TSDs for RSABE/ABEL ElMaestro 2015-05-06 12:31
- TSDs for RSABE/ABEL Helmut 2015-05-06 13:49
- TSDs for RSABE/ABEL - run-time d_labes 2015-05-08 20:47
- Really worth the efforts? Helmut 2015-05-10 01:11
- TSDs for RSABE/ABEL nobody 2015-05-06 11:13