Hi Debbie,

EMA’s GL states “around 50% of the calibration curve range” and FDA’s current draft “in the midrange”
Most colleagues I’ve talked to in recent years spike the middle QC close to \(\sqrt{LLOQ \times ULOQ}\).* That’s justified because calibrators generally follow a geometric progression. Never heard of any problems with assessors/inspectors.



Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮