Calibration range [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2012-09-30 16:15 (4628 d 14:25 ago) – Posting: # 9284
Views: 2,863

Hi Chintamani,

please follow the Policy of the Forum!

❝ Please justify why we take lower limit of quantification (LLQC) 4 to 5 Half life of Cmax and Upper limit of quantification (ULOQ) two times of Cmax.


How can I justify what you do? Which Cmax are you taking into account? The mean? Probably a bad idea.

If a drug shows a low between-subject CV ULOQ = 2× mean Cmax will be too high – you would have to deal with an unnecessary wide calibration range. It’s better to use a narrower range and validate dilution – expecting only a few samples to be out – instead. On the other hand if a drug shows high between-subject CV (e.g., polymorphism) an ULQQ of 2× mean Cmax will be too low by far.

The LLOQ should fulfill two requirements:
  1. ≤5% of the (individual!) Cmax values in order to rule out carry-over and
  2. low enough that AUCt ≥ 80% of AUC.
The latter is not applicable if AUC72 is employed. In a steady state study in a switch-over design with substantial accumulation LLOQ ≤5% Cmax is debatable.

Not only the PK of the drug, but also the biopharmaceutical properties of the formulation (IR vs. MR) and the design (SD vs. MD) drive the calibration range.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,682 registered users;
30 visitors (0 registered, 30 guests [including 20 identified bots]).
Forum time: 06:40 CEST (Europe/Vienna)

Being really good at C++ is like being really good
at using rocks to sharpen sticks.    Thant Tessman

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5