Duplicate LLOQ and ULOQ [Bioanalytics]
Dear Auditor,
What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used, unless it fails (> 20 % deviation from nominal at the LLOQ, 15 % at the ULOQ) or there is a clear analytical reason (instrument failure, chromatographic interference). This should be defined in a SOP. Some people take whichever value gives the highest R2, but I don't like that (I don't consider R2 to be a relevant parameter).
The idea of running the LLOQ and ULOQ samples in duplicate is to avoid having to truncate the curve if one of these samples is failing.
Regards
Ohlbe
What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used, unless it fails (> 20 % deviation from nominal at the LLOQ, 15 % at the ULOQ) or there is a clear analytical reason (instrument failure, chromatographic interference). This should be defined in a SOP. Some people take whichever value gives the highest R2, but I don't like that (I don't consider R2 to be a relevant parameter).
The idea of running the LLOQ and ULOQ samples in duplicate is to avoid having to truncate the curve if one of these samples is failing.
Regards
Ohlbe
—
Regards
Ohlbe
Regards
Ohlbe
Complete thread:
- Calibration Curve auditor 2012-01-20 11:15
- Duplicate LLOQ and ULOQOhlbe 2012-01-20 11:59
- Duplicate LLOQ and ULOQ Helmut 2012-01-20 13:09
- Duplicate LLOQ and ULOQ Ohlbe 2012-01-20 14:26
- Duplicate LLOQ and ULOQ Helmut 2012-01-20 18:02
- Calibration Curve auditor 2012-01-21 05:45
- Calibration Curve Helmut 2012-01-21 12:35
- Calibration Curve Ohlbe 2012-01-21 13:44
- Calibration Curve Helmut 2012-01-22 21:48
- Calibration Curve auditor 2012-01-21 05:45
- Duplicate LLOQ and ULOQ Helmut 2012-01-20 18:02
- Duplicate LLOQ and ULOQ Ohlbe 2012-01-20 14:26
- Duplicate LLOQ and ULOQ Helmut 2012-01-20 13:09
- Duplicate LLOQ and ULOQOhlbe 2012-01-20 11:59