Extrapolation of CC: no way! [Bioanalytics]
❝ Calibration range is: 50 to 1200 ng/ ml
❝ ULOQ QC concentration is: 1215 ng/ ml.
❝
❝ I have object for such kind of estimation of ULOQ QC as it is not within the established CC range under which P&A samples were analyzed.
Agree – that’s really a bad procedure.
Hypothetical situation: accuracy 100%, (im)precision 0% (aka the ‘perfect assay’). High QCs will get back-calculated concentrations > ULOQ = cannot be evaluated = (all!) batches fail.
❝ I need your suggestion regarding this and if possible can you please suggest me any guideline containing such kind of references.
Regulations:
- FDA (2001): The QC samples in duplicate at three concentrations (one near the LLOQ (i.e., ≤3 × LLOQ), one in midrange, and one close to the high end of the range) should be incorporated in each assay run.
- EMA (2011): Within- run accuracy should be determined by analysing in a single run a minimum of 5 samples per level at a minimum of 4 concentration levels which are covering the calibration curve range: the LLOQ, within three times the LLOQ (low QC), around 50% of the calibration curve range (medium QC), and at least at 75% of the upper calibration curve range (high QC).
❝ I have referred the entire available article pertinent to this but I could not get any information for the same.
Do you mean the “Arlington III whitepaper”?
Viswanathan CT, Bansal S, Booth B, DeStefano AJ, Rose MJ, Sailstad J, Shah VP, Skelly JP, Swann PG, and R Weiner
Workshop/Conference Report – Quantitative Bioanalytical Methods Validation and Implementation: Best Practices for Chromatographic and Ligand Binding Assays
The AAPS Journal 9(1), E30-E41 (2007)
online resource
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- Extrapolation of Calibration Curve auditor 2011-12-22 06:21
- ULOQ QC Ohlbe 2011-12-22 14:03
- ULOQ QC (validation!) Helmut 2011-12-22 14:22
- Extrapolation of CC: no way!Helmut 2011-12-22 14:15