calibration curve [Bioanalytics]

posted by ElMaestro  – Denmark, 2010-08-10 03:14 (5798 d 17:38 ago) – Posting: # 5748
Views: 2,910

Hi Ioanam,


❝ Do we need to reanalyze these high values?

Yes (meaning: you need to reanalyse the samples that caused the high values).

❝ How can I evaluate the results in this situation?

Your validation probably includes dilution integrity. You dilute by a certain factor and get a signal that corresponds to somewhere between LLOQ and ULOQ. Then you multiply with the dilution factor.

Pass or fail!
ElMaestro