batch reinjection when QC failed [Bioanalytics]

posted by vish14184 – India, 2010-08-07 22:09 (5804 d 20:10 ago) – Posting: # 5741
Views: 4,008

Dear All,

I would like to know that,

1. In one batch 8 QC is inter spread between unknown sample.
During study sample analysis (analytical batch) found that first two QCs are failed due to Analyte area is decrease (Concluded from, Comparison of Analyte area of Calibration standard and failed two QC samples) then, batch was stopped and batch was re injected from beginning after system stabilization. After re-injection, re injected QCs are passing and batch is within acceptance criteria. Can we accept this batch?
ISTD response found constant throughout batch.

2. In BE study, if number of batch failure in bio analytical analysis is more (more than 10%) then required to file any investigation for same.

Regards,
Vishal nakrani