Internal standard variation [Bioanalytics]

posted by Ohlbe – France, 2010-02-22 13:52 (5971 d 19:53 ago) – Posting: # 4810
Views: 3,010

Dear Kaushal,

As there is no guideline you can use whatever criteria you feel are suitable. Once again, have a look at the variability you have in your pre-study validation (particularly at the full-scale precision and accuracy run, if you have one, and at matrix effects). This will help you to adapt your acceptance criteria for each method, if needed.

Regards
Ohlbe

Regards
Ohlbe