QCs vs. calibrators [Bioanalytics]
❝ Sidenote: As may a single QC. 
Yes, sure

❝ I’m interested in the back-calculated concentrations as well. OK, to be more precise, to check the residuals (haha, is the applied weighting scheme of the curve appropriate as established in validation). I have to back-calculate concentrations first, in order to be able to exclude a point.
OK, agreed, of course. But that's not the primary goal of the study.
❝ Again, would you allow that without justification (in my limited experience FDA does), or would you like to see the same justifications you required for QCs?
Yes, I would accept the exclusion of a calibration sample without any "assignable cause", as the FDA would put it.
❝ I got the impression that you would treat calibrators and QCs differently.
I do. I accept calibrators to be excluded from the calibration curve without an "assignable cause", if they are more than 15 % from nominal (20 % at the LLOQ). The back-calculated concentration should be reported in the analytical report even if excluded (clearly identified as excluded), but I don't mind if it is not included in the calculation of the precision and accuracy of the results at each level of concentration (actually I don't mind if the mean and CV of the back-calculated concentration of the calibrators at each level of concentration are not calculated at all - that's not really relevant). But I do consider that all QCs should be included in the calculations, unless you had a true analytical problem (in which case I would not expect the result to be reported at all).
❝ ❝ The same goes with QC samples…
❝
❝ What do you mean by this? Sounds like an oxymoron to me.
Once again: I treat QCs the same way as unknown samples. Report them all unless you had a clear analytical problem, include them all into calculations, no exclusion from the report, or they won't be representative any more.
❝ P.S.: Interesting thread.
Yes

Best regards
Ohlbe
Regards
Ohlbe
Complete thread:
- outliers test for QC sample mathews 2007-10-11 07:58
- outliers test - but not for QCs! Helmut 2007-10-11 15:31
- Outlier test - but not for QC's yuvrajkatkar 2010-01-19 08:49
- Outlier test - but not for QC's H_Rotter 2010-01-19 12:55
- Outlier tests: forget it! Helmut 2010-01-19 17:36
- Outlier tests: forget it! Ohlbe 2010-01-19 19:00
- Outlier tests: forget it! Helmut 2010-01-19 19:16
- Outlier tests: forget it! Ohlbe 2010-01-19 21:00
- QCs vs. calibrators Helmut 2010-01-19 21:49
- QCs vs. calibrators Ohlbe 2010-01-20 19:01
- QCs vs. calibrators Helmut 2010-01-20 19:26
- QCs vs. calibratorsOhlbe 2010-01-20 19:52
- QCs vs. calibrators Helmut 2010-01-20 19:26
- QCs vs. calibrators Ohlbe 2010-01-20 19:01
- Two thirds of QCs H_Rotter 2010-01-20 16:21
- Two thirds of QCs Ohlbe 2010-01-20 18:49
- Outlier tests: forget it!....what if.... moblak 2010-01-25 10:04
- Outlier tests: forget it!....what if.... Ohlbe 2010-01-25 23:19
- QCs vs. calibrators Helmut 2010-01-19 21:49
- Outlier tests: forget it! Ohlbe 2010-01-19 21:00
- Outlier tests: forget it! Helmut 2010-01-19 19:16
- Outlier tests: forget it! Ohlbe 2010-01-19 19:00
- Outlier test - but not for QC's yuvrajkatkar 2010-01-19 08:49
- outliers test - but not for QCs! Helmut 2007-10-11 15:31