Outlier tests: forget it! [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2010-01-19 20:16 (6002 d 17:41 ago) – Posting: # 4620
Views: 9,416

Dear Ohlbe!

❝ Sorry, but I disagree there. You may exclude a standard sample, but I can't see how you could exclude a QC, unless you have an excellent and documented analytical reason to do it. And when I say documented, I don't mean just writing "sample processing error" in your raw data just because the QC fails, but rather something like no internal standard added, or sample not injected, or chromatographic interference. Your QC result is an experimental result, like it or not, outlier or not.


Disagree. FDA (2001, pages 13 and 16):(my emphases)

That’s exactly my example: one value below –15%, but overall bias and precision within ±15%. Of course the value has to be reported and included in the calculation of acc./prec. I don’t get the rationale why you would allow for exclusion of a calibrator (if not changing the model, according to FDA), but not a QC (both spiked matrix, etc.)

BTW, even the term “outlier” with n=2–3 is crazy.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮