Predose sample without IS? [Bioanalytics]
Dear VP!
Interesting approach. To my knowledge nobody asks for a predose sample without IS - only one sample without IS in the calibration curve.
I think so.
The same reason you would give for any other repeated analysis. If you 'see' no more response at the retention time of the IS, most likely it was an error in the sample preparation.
In case (2) probably you should check your procedures: Watch for a potential contamination in the sample preparation, flush the injector with a solution of more organic modifier than the mobile phase, inject a blank/mobile phase after samples with high concentrations.
❝ In Project sample analysis, we got ISTD area in predose sample though it
❝ is predose sample without ISTD.
Interesting approach. To my knowledge nobody asks for a predose sample without IS - only one sample without IS in the calibration curve.
❝ In that case we have repeated predose sample analysis. Is it correct?
I think so.
❝ If yes, What reason we should mention for that?
The same reason you would give for any other repeated analysis. If you 'see' no more response at the retention time of the IS, most likely it was an error in the sample preparation.
- If the response is roughly the same you would expect for a normal sample (with IS) most likely the analyst erroneously added the full amount of IS.
- If the response was much lower, you may suspect carry-over (both in pipetting or in the autosampler).
In case (2) probably you should check your procedures: Watch for a potential contamination in the sample preparation, flush the injector with a solution of more organic modifier than the mobile phase, inject a blank/mobile phase after samples with high concentrations.
—
Dif-tor heh smusma 🖖🏼 Довге життя Україна!![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- Repeat analysis VP 2009-06-25 14:09
- Predose sample without IS?Helmut 2009-07-06 18:41
- Predose sample without IS? VP 2009-07-21 12:53
- Repeat analysis ElMaestro 2009-07-06 20:11
- Repeat analysis Geokad 2009-07-07 18:35
- Repeat analysis santosh kumar Manthri 2009-07-17 12:30
- Repeat analysis Helmut 2009-07-17 13:26
- Predose sample without IS?Helmut 2009-07-06 18:41