Ruggedness experiment [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2008-12-30 17:57 (6386 d 12:58 ago) – Posting: # 2993
Views: 2,964

Dear Anil!

❝ Can ruggedness expriment results include in intra day or inter day P&A?


USP defines ruggedness as "the degree of reproducibility of test results obtained by the analysis of the same samples under a variety of normal test conditions such as:
  • Different laboratories
  • Different analysts
  • Different instruments
  • Different reagent lots
  • Different analysis days
  • Different elapsed assay times
  • Different assay temperatures..."
Different analysis days are already part of the standard method validation procedures.
For HPLC I would consider the most important factors different instruments, column batches (and/or vendors), slight changes in the composition of the mobile phase and/or the flow rate, and the column temperature. Although FDA's http://www.fda.gov/cder/guidance/4252fnl.pdf guideline (2001) calls for partial validation of "bioanalytical method transfers between [...] analysts" I have never seen a report entitled "Validation of bioanalytical method A for the determination of B in matrix C to be performed by analyst X". :-D
Generally precision of accuracy of QCs are evaluated in ruggedness experiments; for a statistical approach see this article.


Edit: Link corrected for FDA’s new site. [Helmut]

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,571 registered users;
182 visitors (0 registered, 182 guests [including 46 identified bots]).
Forum time: 07:56 CEST (Europe/Vienna)

Ignorance more frequently begets confidence
than does knowledge.    Charles Darwin

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5