Manual re-integration [Bioanalytics]
Dear Najat!
No. The analyst should visually inspect all chromatograms. For a particular chromatogram the automatically changed integration mode may be ‘correct’; in such a case leave it as it is. Only if the automatic integration obviously failed (this may happen quite often in the lower signal range where a lot of noise is present) the analyst should set the baseline manually. Most analysts I know would simply draw the new baseline with the mouse, rather than change integration parameters a couple of times until the system comes up with a better integration (this would result in a lot of trial-and-error…). If you change the integration parameters for the entire batch, you would again have to inspect all chromatograms, and repeat this cycle maybe forever…
❝ ❝ That’s the reason one should always inspect all chromatograms and go with a manual integration if the automatic one obviously failed.
❝
❝ Does this mean that all the chromatograms within a run should have the same integration mode (e.g. base to base) and if any chromatograms were integrated with a different mode (e.g. valley) within the same run they should be manually re-integrated to the (base to base) mode?
No. The analyst should visually inspect all chromatograms. For a particular chromatogram the automatically changed integration mode may be ‘correct’; in such a case leave it as it is. Only if the automatic integration obviously failed (this may happen quite often in the lower signal range where a lot of noise is present) the analyst should set the baseline manually. Most analysts I know would simply draw the new baseline with the mouse, rather than change integration parameters a couple of times until the system comes up with a better integration (this would result in a lot of trial-and-error…). If you change the integration parameters for the entire batch, you would again have to inspect all chromatograms, and repeat this cycle maybe forever…

—
Dif-tor heh smusma 🖖🏼 Довге життя Україна!![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- LC-MS/MS automatic integration najatmughrabi 2008-11-19 07:32
- LC-MS/MS automatic integration Helmut 2008-11-19 12:51
- LC-MS/MS automatic integration najatmughrabi 2008-11-19 14:30
- Manual re-integrationHelmut 2008-11-19 15:35
- LC-MS/MS automatic integration najatmughrabi 2008-11-19 14:30
- LC-MS/MS automatic integration Ohlbe 2008-11-20 00:18
- LC-MS/MS automatic integration Helmut 2008-11-19 12:51