Balance calibration [Bioanalytics]

posted by Ohlbe – France, 2008-11-08 14:50 (6438 d 22:45 ago) – Posting: # 2642
Views: 3,689

Dear Helmut and V. Pardhasaradh,

❝ 3. External weights are also used daily; at least one of two certified

❝ weights (10 g and 50 g - class E2) is used.


If the balance is to be used to prepare stock solutions, where you usually weigh a few mg of reference substance, I would suggest to use control weights in the same range (e.g. 10 mg).

Regards
Ohlbe