Dear V. Pardhasaradhi!
I admit with HS and I have additional information. For Microbalance calibration in Bioanalytical lab E1 or E2 weights can be used. Minimum weight cloud be anywhere between 0.5µg-2mg. That's why you can calibrate internally with 2mg and max 1g. Additionally you can also get the balance calibrated with external agencies and calculate measurement uncertainty at defined frequency.
Regards
I admit with HS and I have additional information. For Microbalance calibration in Bioanalytical lab E1 or E2 weights can be used. Minimum weight cloud be anywhere between 0.5µg-2mg. That's why you can calibrate internally with 2mg and max 1g. Additionally you can also get the balance calibrated with external agencies and calculate measurement uncertainty at defined frequency.
Regards
—
Best Regards
Joshi
Best Regards
Joshi