Representation of BQL values in concentration data [Bioanalytics]

posted by ElMaestro  – Denmark, 2019-02-21 07:58 (2284 d 17:20 ago) – Posting: # 19953
Views: 5,098

Hello GM,

❝ As per the new Bioanalytical Method validation guidelines, "Study samples with concentrations listed below the LLOQ should be reported as below the LLOQ (BQL). Study samples with concentrations above the ULOQ should be diluted and re-analyzed, or the standard curve should be extended and revalidated."


❝ This means, whether we have to replace BQL values with "zero" or can we keep as it is in the BE analysis without replacing those values with zero?


Those values are not measured as zero, they are BLLOQ or BQL or BLQ or whatever code you assign to them. However, it is a commonly accepted principle that when you do NCA on the series, you consider BLQ to be zero for the purpose of derivation of AUC. This works without problems generally, with a few exceptions; for example, there is no universal agreement on how to handle a BLQ between quantifiable concentrations after tmax when you do "linear up, log down". But I assure you there are very strong personal opinions :-D

Your processes around sampling and reporting PK-data and doing NCA should be in an SOP, in my opinion.

❝ We have a situation that, one of our study is passing the BE with BQL values (without replacing those values with zero), but failed to attain the BE without BQL values(with replacing those values with zero).

Sounds like you are in a good position. But check what the software actually does with BLQs - at the end of the day this is what determines if the proof of BE is rigorous and scientific.

Pass or fail!
ElMaestro

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,675 registered users;
16 visitors (0 registered, 16 guests [including 10 identified bots]).
Forum time: 02:19 CEST (Europe/Vienna)

An expert is someone who knows some of the worst mistakes
that can be made in his subject,
and how to avoid them.    Werner Heisenberg

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5