Metabolite back conversion [Bioanalytics]

posted by Ohlbe – France, 2017-07-18 18:29 (2868 d 02:47 ago) – Posting: # 17559
Views: 5,577

Dear Samar Sameh,

❝ We are dealing with 2 metabolites both on the +ve ionization mode; the N-oxide (potency=96.8%) and sulfoxide (potency=99.9%)


I have no experience with sulfoxide. N-oxides are known for in-source dissociation. My understanding is that source temperature is an important factor. However if the metabolite is chromatographically separated from the parent, in-source dissociation should not create any problem.

If you are dealing with 2 metabolites and are finding potential issues, I would suggest to study the two separately in order to facilitate troubleshooting.

Regards
Ohlbe