❝ According to my boss, its "common sense" that the conc. shouldn't be same.


Hm, I took a different message home from this thread.

❝ Total CC points was eight so I think we can reprocess MV data with the problematic point removed and the curve will still have enough CC levels.


Actually, I even had a short discussion with a few (OK, 2, but both are no newbies) of my co-workers and one of them had an argument pro different concentrations (basically "I see this all the time") in case of clinical (!) laboratories.
But both (and me) would discourage omission of valid data from the calibration without proper justification (read: would start a discussion on re-calculating/re-validating including all CCs if we saw that in a bioanalytical/validation report).
Just my thoughts.