dead volume definition [Bioanalytics]
Dear Vishal!
I think that you missed one pair of perentheses; the formula should read:
k'= (tR - t0)/t0
One should distinguish between theoretical dead volume (the interstitial volume between the particles of the package material + the porous volume of the packing material) and the experimental dead volume, which includes the volume of the entire LC system (injector, mixing chamber, fittings, connecting capillaries, etc).
For a glossary of terms see this article from LCGC Europe (Feb 2002).
The one compound we used in getting an experimental value for t0 in RP LC is thiourea - SC(NH2)2; others are uracil and acetone.
You may calculate the dead volume (VM) from the geometry of your column (length, diameter), the flow rate and the retention time.
VM = t0 × v, where v=velocity of the mobile phase.
An approximate formula uses the column's inner diameter d [cm] and column length L [cm], as well as the flow rate F [ml/min]:
VM = 0.5 × d2 × L, and tM = VM / F
Example:
d=0.46cm, L=15cm, F=1.5mL/min
VM=1.6mL, tM=1.1min
You will hardly find a compound with a tR < t0 (besides dissolved oxygen...).
In practical LC applications I would not go with a method with a k' of less than 3-5. The higher the k' the better the separation, but you are loosing sensitivity (peaks are getting broader). Therefore k' >10-15 is also not a good idea.
❝ In column specification we have a calcuation for capacity facotor K'
❝ (formula is k'=tr-to/to) and as a 'to', first retained peak of the
❝ combination is consider.
I think that you missed one pair of perentheses; the formula should read:
k'= (tR - t0)/t0
❝ kindly give me the actul procedure to calculate the dead volume
❝ and also discribe how it is effect my chromatogram if the
❝ value of k' is >10, < 1 and >30 ?
One should distinguish between theoretical dead volume (the interstitial volume between the particles of the package material + the porous volume of the packing material) and the experimental dead volume, which includes the volume of the entire LC system (injector, mixing chamber, fittings, connecting capillaries, etc).
For a glossary of terms see this article from LCGC Europe (Feb 2002).
The one compound we used in getting an experimental value for t0 in RP LC is thiourea - SC(NH2)2; others are uracil and acetone.
You may calculate the dead volume (VM) from the geometry of your column (length, diameter), the flow rate and the retention time.
VM = t0 × v, where v=velocity of the mobile phase.
An approximate formula uses the column's inner diameter d [cm] and column length L [cm], as well as the flow rate F [ml/min]:
VM = 0.5 × d2 × L, and tM = VM / F
Example:
d=0.46cm, L=15cm, F=1.5mL/min
VM=1.6mL, tM=1.1min
You will hardly find a compound with a tR < t0 (besides dissolved oxygen...).
In practical LC applications I would not go with a method with a k' of less than 3-5. The higher the k' the better the separation, but you are loosing sensitivity (peaks are getting broader). Therefore k' >10-15 is also not a good idea.
—
Dif-tor heh smusma 🖖🏼 Довге життя Україна!![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- dead volume definition vishal vayeda 2008-01-01 07:56
- dead volume definitionHelmut 2008-01-01 17:02