Method Validation [Bioanalytics]

posted by Ohlbe – France, 2007-12-15 13:05 (6773 d 00:45 ago) – Posting: # 1378
Views: 5,570

Dear Chirag,

No guideline jumps to my mind on this topic. What I would suggest would be to prepare and analyse 3 different QC samples, each spiked at a low concentration (LLOQ or LQC) with one analyte (1, 2 or 3) and ULOQ with the other two. If there is any interference (either directly visible, or influencing the ionisation of your analyte or IS if you are using MS/MS) you would have maximum chances to see an impact on the accuracy.

Regards
Ohlbe