~geometric mean of range [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2014-06-17 16:30 (3998 d 07:51 ago) – Posting: # 13074
Views: 14,516

Hi Debbie,

EMA’s GL states “around 50% of the calibration curve range” and FDA’s current draft “in the midrange”
Most colleagues I’ve talked to in recent years spike the middle QC close to \(\sqrt{LLOQ \times ULOQ}\).* That’s justified because calibrators generally follow a geometric progression. Never heard of any problems with assessors/inspectors.