Plasma separation [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2007-10-19 15:15 (6828 d 00:45 ago) – Posting: # 1224
Views: 5,583

Dear Chirag!

❝ The concept we follow is after collection of blood within 30 minutes separation and storage should be done,…


Some remarks additionally to Ohlbe’s post
It’s quite difficult to validate the degradation process. It would consist of spiking whole blood at 37 ℃ and carrying it through different conditions to the frozen plasma.
Most difficulties arise from:Many people just skip this step in validation (to my knowledge it’s not covered in a guideline) assuming similar degradation for test/reference in BE.
To be on the safe side you may transfer your samples immediately after collection to a thermostated water bath (if you don’t have one, a water tank with ice-cubes or a beaker filled with crushed ice also does the job) until centrifugation. According to the Arrhenius-equation this procedure is expected to bring down the degradation to about 20%.

BTW, how do you actually keep the interval from collection to storage at ≤30 minutes? The run-time of the centrifuge is probably 15-20 minutes and especially in the early collection periods (let’s say the first 2 hours) you must perform the separation in some kind of batch mode – unless you have an entire battery of cooled centrifuges in your department…

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,656 posts in 4,994 threads, 1,570 registered users;
298 visitors (0 registered, 298 guests [including 17 identified bots]).
Forum time: 16:01 CEST (Europe/Vienna)

Most scientists today are devoid of ideas, full of fear, intent on
producing some paltry result so that they can add to the flood
of inane papers that now constitutes “scientific progress”
in many areas.    Paul Feyerabend

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5