Confusing table [Bioanalytics]
Dear Helmut,
Thank you very much for your explanation.
May I know how the values of 1/x and 1/x2 are generated?
Please kindly show how the weights are generated.
We use the weight to multiply the independent values. Then, plot standard curve. From the curve, we can estimate the slope and intercept. We can then compute the back calculated value. The flow is right
??
Appreciate your input.
Thank you.
Regards,
Ken
Thank you very much for your explanation.
May I know how the values of 1/x and 1/x2 are generated?
❝ x is the independent variable (in bioanalytics concentration) and y the dependent (observed) one – in BA the response (analyte/IS-ratio, peak height/area, whatsoever). The weights are not given in the table, they are simply 1/x or 1/2.
Please kindly show how the weights are generated.
We use the weight to multiply the independent values. Then, plot standard curve. From the curve, we can estimate the slope and intercept. We can then compute the back calculated value. The flow is right
??Appreciate your input.
Thank you.
Regards,
Ken
Complete thread:
- Standard calibration curve Ken Peh 2014-01-13 17:20
- Standard calibration curve Ohlbe 2014-01-13 21:50
- Use of weighting factor Ken Peh 2014-01-14 16:37
- Confusing table Helmut 2014-01-14 17:32
- Use of weighting factor Ken Peh 2014-01-14 16:37
- Standard calibration curve ElMaestro 2014-01-14 10:14
- Standard calibration curve Roberto 2014-02-18 13:27
- Standard calibration curve Ohlbe 2014-02-18 14:53
- Standard calibration curve Roberto 2014-02-18 17:00
- Standard calibration curve Ohlbe 2014-02-18 14:53
- Standard calibration curve Roberto 2014-02-18 13:27
- Standard calibration curve AngusMcLean 2014-03-01 01:27
- Standard calibration curve Ohlbe 2014-01-13 21:50