Dear Members,

In EMEA Guideline on bioanalytical method validation, under 4.1.4 Standard Calibration Curve, it is written that "The back calculated concentrations of the calibration standards should be within ±15% of the nominal value, except for the LLOQ for which it should be within ±20%"

My questions are as below:
  1. Besides the accuracy values, is regression coefficient value important, eg. R square >0.99 ? Should both be considered as acceptance criteria? The accuracy and R square values.
  2. If we have 8 calibrators, from back calculated concentration of the calibration standards, we found 2 calibration standards failed accuracy. One of calibration standard was LLOQ. To keep the LLOQ, instead of dropping the failed LLOQ, reject another calibration standard which passes the accuracy and reevaluate the calibration curve. As a result, 7 calibration standards pass. Is this the right approach :confused:??
  3. On what occasion weighting method is allowed ? Weighting method is not mentioned in EMEA guideline but in FDA new draft guideline.
Highly appreciate your kind comments.

Thank you.

Regards,
Ken

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,674 registered users;
66 visitors (0 registered, 66 guests [including 8 identified bots]).
Forum time: 10:35 CEST (Europe/Vienna)

Every man gets a narrower and narrower field of knowledge
in which he must be an expert in order to compete with other people.
The specialist knows more and more about less and less
and finally knows everything about nothing.    Konrad Lorenz

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5