Calculation of ISR deviation [Bioanalytics]

posted by Roberto – Italy, 2013-12-09 13:52 (4188 d 09:25 ago) – Posting: # 12026
Views: 7,189

Hi ElMaestro,

What do you think about my interpretation here below described?

I do not think that EMA provides broader limits for ISR than those widely knew and shared.
I believe that what is written in the BIOANALYTICAL GUIDELINE ON METHOD VALIDATION of 21 July 2011, it's just bad written.

If we start from the DRAFT 19 November 2009, at paragraph 6. Incurred Samples Reanalysis, at line 485 is reported: "The difference between the two values ​​obtained should be within 20% of the mean (30% for ligand-binding assays) for at least 67% of the repeats."

Continuing in the document "Overview of comments received on 'Guideline on validation of bioanalytical methods' (EMEA/CHMP/EWP/192217/2009)", the concept widely reported in the WPs is mentioned and i would highlight a couple of points:

Line no. 467-477
Stakeholder no. 24
Comment and rationale;proposed changes: Chapter should be adapted according to EBF recommendations. as described in “Incurred Sample Reproducibility: Views and Recommendations by the European Bioanalysis Forum, P. Timmerman et al” (PDF attached)
Outcome: It is considered that the revised paragraph covers sufficiently the recommendations of the EBF.
Regarding the criteria, 20% difference from the mean would be in line with the White paper (Fast et all., AAPS Journal, DOI 10.128/s12248-009-9100-9)

Line no. 467-486
Stakeholder no. 39
Comment and rationale;proposed changes: … And finally, the requirement for the difference of two measurements to be within 20% is more strict than that for QC samples.
Outcome:There was an international consensus on this criteria (see White Paper).

Finally, in the document "Appendix IV of the Guideline on the Investigation on Bioequivalence (CPMP/EWP/QWP/1401/98 Rev.1): Presentation of Biopharmaceutical and Bioanalytical Data in Module 2.7.1" at the point "Table 4.3 Sample analysis of <Study ID> ", under the heading Incurrent sample reanalysis, it is repeated the same acceptance criterion "Percentage of samples where the difference between the two values ​​was less than 20% of the mean for chromatographic assays or less than 30% for ligand binding assays. "

In conclusion, I believe that the ISR acceptance criteria for EMA is that well known and internationally recognized:

Variability (%)= (Repeat – Original) × 100/Mean


and that what is reported in the current GL is written in a not clear way and misleading.

What is your opinion on this?

Thanks at all and best regards

Roberto

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,675 registered users;
68 visitors (0 registered, 68 guests [including 13 identified bots]).
Forum time: 00:18 CEST (Europe/Vienna)

Every man gets a narrower and narrower field of knowledge
in which he must be an expert in order to compete with other people.
The specialist knows more and more about less and less
and finally knows everything about nothing.    Konrad Lorenz

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5