Choose and Pick in method validation [Bioanalytics]

posted by Ohlbe – France, 2013-11-26 19:45 (4201 d 08:46 ago) – Posting: # 11978
Views: 3,597

Dear Ken,

If your run still passes for precision and accuracy: just report it, everything is fine. Nobody ever said that each and every individual value in the run had to pass.

If the run fails for precision and/or accuracy: repeat it, but report the failed run in your validation report too. These will be considered as too independent runs (no pooling !). The FDA and EMA will ask you to calculate the global precision and accuracy with all runs, whether passing or failing.

If another run fails: revise your method.

Regards
Ohlbe