Reporting QCs [Bioanalytics]
dear ohlbe
I value your feed back notes
what is the minimum re-analysis points is accepted to demonstrate reproducibility from the whole incurred samples?
re-analysis done in duplicate are enough?
what if my plasma sample volumes does not allow me to do re-analysis for more than twice?
Regards
Charl
I value your feed back notes
❝ The FDA wants you to demonstrate the precision of your method using real-life samples, not only spiked samples. Meaning that if you want to comply with this requirement you will have to re-assay some of the samples from your study. The idea is to use the results of the first determination for your PK analysis, and have a table somewhere listing the results of whatever you will have done to demonstrate precision on incurred samples.
❝ You do not need to treat these results as per your procedure on repeat analysis to select which result should be reported.
what is the minimum re-analysis points is accepted to demonstrate reproducibility from the whole incurred samples?
re-analysis done in duplicate are enough?
what if my plasma sample volumes does not allow me to do re-analysis for more than twice?
Regards
Charl
Complete thread:
- Reporting Failed QCs Dr. Harish L. Rao 2007-08-30 08:55
- Reporting QCs Helmut 2007-08-30 12:05
- Reporting QCs Ohlbe 2007-08-31 10:46
- Reporting QCs Charl 2007-09-02 15:11
- Reporting QCs Ohlbe 2007-09-02 23:50
- Reporting QCsCharl 2007-09-03 10:13
- Precision of incurred samples Ohlbe 2007-09-03 10:20
- Problematic incurred samples guidelines Helmut 2007-09-03 19:49
- Precision of incurred samples Ohlbe 2007-09-03 10:20
- Reporting QCsCharl 2007-09-03 10:13
- Reporting QCs Ohlbe 2007-09-02 23:50
- Reporting QCs Charl 2007-09-02 15:11
- Reporting QCs Ohlbe 2007-08-31 10:46
- Reporting QCs Helmut 2007-08-30 12:05