Reporting QCs [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2007-08-30 14:05 (6872 d 04:10 ago) – Posting: # 1041
Views: 8,846

Dear Harish!

❝ Let's first assume that we are not performing any outlier test.


Good idea - would be possible for n>6 anyway, which in validation generally is performed only for LLOQ-samples.

❝ During Bioanalytical Method Validation, do we need to include the failed

❝ QCs in the Inter & Intra batch summary table for Mean & CV%

❝ calculations?


Yes, you should include all measured values in the report (your method's specifications should be transparent).
Excluded values should be flagged (i.e., by means of an asterisk leading to a footnote, or striken through).
I would suggest to present two tables for Accuracy/Precision; one with and the other one without excluded values - making the life of auditors and regulatory inspectors less miserable.

See also this thread.