Difference in values between reanalysed and ISR samples [Bioanalytics]

posted by Ken Peh – Malaysia, 2013-03-16 15:40 (4457 d 22:06 ago) – Posting: # 10206
Views: 4,007

Dear Cakhatri,

According to EMEA Guideline on Bioanalytical Method Validation (2011), the concentration obtained for the initial analysis and reanalysis should be within 20% of their mean for at least 67% of the repeats. Large differences between results may indicate analytical issues and should be investigated.

If I interpret the guideline correctly, as long as you have more than 67% of the repeats within 20% of their means, you are safe.

What do you mean when you mentioned "ISR passed with more than 95% acceptance criteria" ?

I have no experience dealing with FDA on ISR. Any guideline by FDA on this matter ??

Regards,
Ken