Difference in values between reanalysed and ISR samples [Bioanalytics]

posted by Ohlbe – France, 2013-03-11 19:58 (4854 d 23:00 ago) – Posting: # 10188
Views: 5,281

Dear Chirag,

❝ When the initial and repeat data of drug was compared, the difference observed was about +/- 30%. This was the case with 3 subjects out of more than 65 subjects


❝ ISR passed with more than 95% acceptance criteria


❝ Does this cause concern where the difference between initial and repeat analysis data is more than 30% and ISR is still meeting the acceptance criteria.


Some questions:

- in how many runs were the samples of these 3 subjects analysed initially ? How many runs for re-analysis ?

- was the difference always in the same direction for all samples for each subject ? If yes, it just looks like an accuracy issue (one run may be 15 % too high, the second 15 % too low).

- was the difference in the same direction for the 3 subjects ? Higher, or lower than the initial result ? If the difference was systematically in the same direction, with 3 independent runs, you may have an issue with stability (either of the parent analyte, or of a metabolite). Did you analyse your ISR runs only at the end of the study, or on an ongoing basis each time you had enough samples to re-analyse in a run ?

Regards
Ohlbe

Regards
Ohlbe

Complete thread:

UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
325 visitors (0 registered, 325 guests [including 21 identified bots]).
Forum time: 19:59 CEST (Europe/Vienna)

Ignorance more frequently begets confidence
than does knowledge.    Charles Darwin

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5