Removal of data points from CC [Regulatives / Guidelines]

posted by Helmut Homepage – Vienna, Austria, 2007-10-15 21:30 (6832 d 17:21 ago) – Posting: # 1192
Views: 6,345

Dear Ramesh - or Debbie in PharmPK ;-)

❝ … all the standards (STD 1 to STD 11) are with in the specifications for accuracy, but, both blank and blank IS are not meeting the acceptance criteria (less 20 % of the LOQ(STD 1)) against the STD 1.


Where does this criterion (<20% of LLOQ) come from?
And how do you calculate it actually - since you don't know the concentration, do you compare the measured response of the blank with the calculated response at zero (=intercept of the CC)?

❝ so we have removed the STD 1 from the regression and calculated the interference in the STD balnk and STD Zero samples (blank sample with IS), against the STD 2, making STD 2 as the new LOQ, …


… in trusting the interference more than your calibrators?

Most people would allow removal of data points from the calibration curve only (for <25% of data points) if the back-calculated concentration deviates more than 15% (or 20% at the LLOQ) from the nominal concentration in the following cases:Removal is possible only for CC data points back-calculated from the entire regression model. IMHO you are trying to remove data points, because you have some problems with contamination in both CC-blanks.
I assume, you don’t have contaminations at the other levels of the CC, because in this case you would get an increased (compared to other CCs) intercept, which should give a back-calculated value of ≪LLOQ for the blanks.

Your method of defining a new LLOQ higher than than the lowest QC-level simply leads to a rejected run.


Edit: Link corrected for FDA’s new site. [Helmut]

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,656 posts in 4,994 threads, 1,570 registered users;
148 visitors (0 registered, 148 guests [including 24 identified bots]).
Forum time: 14:51 CEST (Europe/Vienna)

It requires a very unusual mind
to undertake the analysis of the obvious.    Alfred North Whitehead

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5