Suggested code [Two-Stage / GS Designs]

posted by ElMaestro  – Denmark, 2017-08-21 20:13 (3221 d 20:38 ago) – Posting: # 17729
Views: 33,502

Hi Hötzi,


I imagine I was not able to explain what I meant??
I looked at the code available here.


Do this:

1. In the R code for the Potvin function, include this code:

Power.calc=function(Nps, GMR, CV, Is.St2=0)
{
  s=sqrt(log(1+CV*CV))
  nom1=log(1.25/GMR)
  nom2=-log(1.25*GMR)
  den=s*sqrt(2/(2*Nps))
  df=Nps*2-2
  if (Is.St2==T) df=df-1
  ##cat("1-a=", 1-alpha, "df=",df,"\n")
  if (Is.St2==0) q=vecQT1[df]
    else         q=vecQT2[df]
  pw=pt( (nom1/den) -q, df) - pt( (nom2/den) +q, df)
  if (pw<0) pw=0.01
  return(pw)
}

FindCritCV=function(Pwr.T, GMR, Nps, Is.St2=1, toler=0.000001)
{
  CV1=0.0123
  CV2=8.7654
  while (abs(
         Power.calc(Nps, GMR, CV1, Is.St2) -
         Power.calc(Nps, GMR, CV2, Is.St2))>toler)
  {
   CV3=0.5*(CV1+CV2)
   tmp=Pwr.T-Power.calc(Nps, GMR, CV3, Is.St2)

   if (tmp>0) CV2=CV3
      else    CV1=CV3
  }
  return(0.5*(CV2+CV1) )
}



2. immediately before the line that says "if (setseed) set.seed(1234567)", include this code:
MAX=1000 ##Hötzi, specify the number of values you need,
         ##I don't know it from the top of my head.
vecQT1<<-c(1:MAX)
vecQT2<<-c(1:MAX)
for (i in 1:MAX)
   {
    vecQT1[i]<<-qt(p=1-alpha[1], df=i)
    vecQT2[i]<<-qt(p=1-alpha[2], df=i)
   }


vecCritCVs<<-rep(0,MAX) 
for (Nps in 6:length(vecCritCVs))
  vecCritCVs[Nps]<<-FindCritCV(0.8, 0.95, Nps, 1, 1e-10)
  ##you can tweak this line with user setting for
  ##assumed GMR and target power as you please,
  ##using the values from the function call


3. Replace the code for SampleN.Stg2 with the code from my previous post:

sampleN.Stg2.EM <- function(CV)

❝ {

❝   #all other variable info is already accumulated thin the critical values, so:
❝   Nps=1
❝   while (vecCritCVs[Nps]<=CV) Nps=Nps+1
❝   return(Nps*2) 

❝ }

(and make sure you call this function, possibly rename it as you deem appropriate)

4. Tweak and benchmark it. The relevant benchmark is the runtime for the entire Potvin call with 1.000.000 iterations at e.g. CV=0.3 and N1=36 total or something which takes tangible computation time. You will also need to modify the Power.calc function to accomodate other acceptance ranges.

5. Eliminate all other qt calls by replacing them with lookups into the vecQT1 or vecQT2 vectors.



Note: I did not test it, I only looked at the code online, so not everything here can be assumed to be working out of the box.

Pass or fail!
ElMaestro

Complete thread:

UA Flag
Activity
 Admin contact
23,654 posts in 4,992 threads, 1,571 registered users;
139 visitors (0 registered, 139 guests [including 8 identified bots]).
Forum time: 16:51 CEST (Europe/Vienna)

Scientists often have a naïve faith that
if only they could discover enough facts about a problem,
these facts would somehow arrange themselves
in a compelling and true solution.    Theodosius Dobzhansky

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5