Winnonlin Bug [Software]

posted by Dr. Harish L. Rao – India, 2009-03-13 05:18 (5889 d 06:48 ago) – Posting: # 3357
Views: 3,821

(edited on 2009-03-13 05:47)

Dear Group Members,

I would like to draw your attention on a potential bug found in Winnonlin 5.2.1 software. Here are the details:

During Winnonlin PK data analysis, Lambda_z range needs to be selected. This could be done by one of the following methods:

1. Setting up the Start & End times (terminal phase) or,

2. Manually deselecting points one by one (by holding Ctrl key while selecting or deselecting a point) from a set of all points.

Although there were no differences between the two methods in Lambda_z, Half Life, Rsq, Cmax, Tmax, Tlast & Clast values, significant differences were observed in C0 & its related parameters such as, Vz, Vss, AUCINF, MRT, etc. It appears that WinNonlin is doing the log-linear regression with the first two points when no exclusions are applied; however, it is defaulting to the rule of using the first non-zero concentration for C0 when the first two points are excluded.

Pharsight Corporation admits that this is a bug & states that a potential fix would be developed in the next release. Until then, it is recommended to use the first option (selection of Start & End times of terminal phase).

Best regards,
Dr. Harish L. Rao
India.

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,667 registered users;
25 visitors (0 registered, 25 guests [including 8 identified bots]).
Forum time: 13:06 CEST (Europe/Vienna)

It is true that many scientists are not philosophically minded
and have hitherto shown much skill and ingenuity
but little wisdom.    Max Born

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5