auditor
★    

India,
2012-01-20 12:15
(5256 d 21:49 ago)

Posting: # 7973
Views: 6,777
 

 Calibration Curve [Bioanalytics]

Hi,
We would like to know how to plot calibration curve if there are duplicate standards used at LLOQ & ULOQ levels?
For example: if we take 8 standards (each of diferent level) for plotting a curve, wherein 2 levels are in duplicate (LLOQ & ULOQ) i.e. total 10 no of samples (standards). In such case if all the standards are paasing the criteria including both the duplicates then all 10 standard needs to be considered or mean of duplicates (LLOQ & ULOQ) needs to be considered while plotting the curve?

Regards,
Auditor
Ohlbe
★★★

France,
2012-01-20 12:59
(5256 d 21:05 ago)

@ auditor
Posting: # 7975
Views: 6,139
 

 Duplicate LLOQ and ULOQ

Dear Auditor,

What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used, unless it fails (> 20 % deviation from nominal at the LLOQ, 15 % at the ULOQ) or there is a clear analytical reason (instrument failure, chromatographic interference). This should be defined in a SOP. Some people take whichever value gives the highest R2, but I don't like that (I don't consider R2 to be a relevant parameter).

The idea of running the LLOQ and ULOQ samples in duplicate is to avoid having to truncate the curve if one of these samples is failing.

Regards
Ohlbe

Regards
Ohlbe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-01-20 14:09
(5256 d 19:55 ago)

@ Ohlbe
Posting: # 7980
Views: 6,156
 

 Duplicate LLOQ and ULOQ

Dear Ohlbe!

❝ What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used,…


Really most people? That’s strange to me. Why the first and not the second? Why not both?

❝ Some people take whichever value gives the highest R2, but I don't like that (I don't consider R2 to be a relevant parameter).


Right. R² is meaningless in calibration (if you prefer the statistical term: inverse regression). A nice reference (based on backcalculated concentrations):

Almeida AM, Castel-Branco MM, and AC Falcão
Linear regression for calibration lines revisited: weighting schemes for bioanalytical methods
Journal of Chromatography B 774, 215–22 (2002)


Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Ohlbe
★★★

France,
2012-01-20 15:26
(5256 d 18:38 ago)

@ Helmut
Posting: # 7982
Views: 6,040
 

 Duplicate LLOQ and ULOQ

Dear Helmut,

❝ ❝ What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used,…


❝ Really most people?


In my experience, yes. I mean, most people analysing only the first and last samples in duplicate.

❝ That’s strange to me. Why the first and not the second? Why not both?


Their rationale is only to avoid having to truncate the curve if the first or last samples fail. If there is no objective reason to use the second value, they use the first to avoid nasty discussions with regulators (such as: the run passes with the second injection but fails with the first, why did you choose the second ? Did the first one fail by accident, or did the second one pass by chance ?).

Using both repetitions would make sense to improve the fit (the two ends of the curve are where you have the more uncertainty). But then you get into other discussions such as: if you consider you need duplicate analysis to get a correct fit, then you should also analyse the subject samples in duplicate...

Some other labs analyse the whole curve in duplicate, once at the start of the run and once at the end, to compensate for possible drifts. Personally if there is a drift I would rather shorten the run... But I think that's a discussion we had previously on the forum.

Regards
Ohlbe

Regards
Ohlbe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-01-20 19:02
(5256 d 15:02 ago)

@ Ohlbe
Posting: # 7983
Views: 6,084
 

 Duplicate LLOQ and ULOQ

Dear Ohlbe!

❝ ❝ ❝ What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used,…

❝ ❝

❝ ❝ Really most people?


❝ In my experience, yes. I mean, most people analysing only the first and last samples in duplicate.


Ah, now I get your point!

❝ […] if you consider you need duplicate analysis to get a correct fit, then you should also analyse the subject samples in duplicate...


From a statistical point of view the best weighting scheme would be 1/s² (1/x or 1/x² are crude attempts to deal with multiplicative error). I don’t want to start a discussion whether it makes sense to calculate s² from n=2. :-D But there is a drawback: 1/s² can only be used for pretty ‘stable’ methods where one would never drop any calibrator.

❝ Some other labs analyse the whole curve in duplicate, once at the start of the run and once at the end, to compensate for possible drifts. Personally if there is a drift I would rather shorten the run... But I think that's a discussion we had previously on the forum.


Yes…

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
auditor
★    

India,
2012-01-21 06:45
(5256 d 03:19 ago)

@ Helmut
Posting: # 7984
Views: 5,982
 

 Calibration Curve

Thanks for the comments.
However, we still not get the precise response for above said concern i.e. if we are taking duplicates (LLOQ & ULOQ) then we should consider both the duplicates (10 no. of samples) or mean of duplicates (LLOQ & ULOQ) for plotting?
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-01-21 13:35
(5255 d 20:29 ago)

@ auditor
Posting: # 7985
Views: 5,993
 

 Calibration Curve

Dear Auditor!

❝ […] if we are taking duplicates (LLOQ & ULOQ) then we should consider both the duplicates (10 no. of samples) or mean of duplicates (LLOQ & ULOQ) for plotting?


Don’t plot means of duplicates (loss of visual information about their variability). Plot all data points you have used in the calibration, regardless whether you have used all n=10, dropped one (n=9), or two (n=8). Overlay the calibration curve based on the estimated slope and intercept.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Ohlbe
★★★

France,
2012-01-21 14:44
(5255 d 19:21 ago)

@ auditor
Posting: # 7986
Views: 5,935
 

 Calibration Curve

Dear Auditor,

Sorry I had misunderstood your question, I thought you were asking which point to use for calculations. Regarding plotting usually HPLC or LC-MS/MS software can plot all points, whether used or not. Points excluded from the calculations will look differently (e.g. clear circles instead of black dots). I would plot all points, including unused duplicates. Don't plot means, unless you are using means for calculations.

Regards
Ohlbe

Regards
Ohlbe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-01-22 22:48
(5254 d 11:16 ago)

@ Ohlbe
Posting: # 7988
Views: 6,028
 

 Calibration Curve

Dear Ohlbe!

❝ […] usually HPLC or LC-MS/MS software can plot all points, whether used or not. Points excluded from the calculations will look differently (e.g. clear circles instead of black dots). I would plot all points, including unused duplicates.


Hey, that’s even better!

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,653 posts in 4,991 threads, 1,571 registered users;
135 visitors (0 registered, 135 guests [including 21 identified bots]).
Forum time: 11:05 CEST (Europe/Vienna)

Outside his own ever-narrowing field of specialization,
a scientist is a layman.
What members of an academy of science have in common
is a certain form of semiparasitic living.    Erwin Chargaff

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5