2020-03-11 08:01
(197 d 07:12 ago)

Posting: # 21237
Views: 1,458
 

 Rejection of the standard curve [Bioanalytics]

Dear Colleagues
I need your opinions about the following:
we prepare the standard curve samples by using a human plasma lot, this curve composed from 8 non zero points, one blank sample and one blank sample with adding internal standard, the samples ordered to start with the blank then the blank with IS and then the non standard zero pints.
Upon analysis in the LC-MS/MS system, the blank sample with the IS found contains an interfering peak at the analyte retention time with more than 20% from the LLOQ response, while the blank sample is clean and the LLOQ response is normal (when compared to the previously analyzed runs). We conclude that the blank with IS sample was contaminated during handling while the other samples are not contaminated.

My question: is this case an enough reason to reject the standard curve and then to reject the run results? or can we accept the run results?

Thanks in advance for your thoughts.


Edit: Category changed; see also this post #1[Helmut]
ElMaestro
★★★

Belgium?,
2020-03-11 08:55
(197 d 06:18 ago)

@ Ayman Rabayah
Posting: # 21238
Views: 596
 

 Rejection of the standard curve

Hello Ayman,

» We conclude that the blank with IS sample was contaminated during handling while the other samples are not contaminated.

Or it could be crosstalk. Or your deuterated IS (if that is what you have) is not very pure but contains interfering amounts of analyte.

I could be wrong, but...

Best regards,
ElMaestro

R's base package has 274 reserved words and operators, along with 1761 functions. I can use 18 of them (about 14 of them properly). I believe this makes me the Donald Trump of programming.
Activity
 Admin contact
21,074 posts in 4,394 threads, 1,468 registered users;
online 8 (2 registered, 6 guests [including 2 identified bots]).
Forum time: Thursday 16:14 CEST (Europe/Vienna)

Enlightenment is man’s emergence from his self-imposed nonage
for which he himself was responsible.
Nonage is the inability to use one’s own intellect
without the direction of another.    Immanuel Kant

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5