Mutasim ☆ Jordan, 2019-08-07 15:16 (1717 d 08:09 ago) Posting: # 20484 Views: 9,125 |
|
Dear Members, What is the explanation for and the impact of having a significant group by sequence interaction in a study with 4 groups and 2 sequences and 4 periods (HVD). |
Helmut ★★★ Vienna, Austria, 2019-08-07 16:35 (1717 d 06:50 ago) @ Mutasim Posting: # 20485 Views: 7,911 |
|
Salam Mutasim, ❝ What is the explanation for … Chance. ❝ … and the impact of having a significant group by sequence interaction in a study with 4 groups and 2 sequences and 4 periods (HVD). If you are dealing with ABEL (the EMA’s method) chances than any (!) European agency will ask for a group-term in the analysis are close to nil. When I gave my first presentation about it in Budapest 2017 in front of European regulatory (!) statisticians they were asking in disbelieve “What the hell are you talking about?” See what the BE-GL states: The study should be designed in such a way that the formulation effect can be distinguished from other effects. This is in line what is stated in the Q&A document about Two-Stage Designs: A model which also includes a term for a formulation*stage interaction would give equal weight to the two stages, even if the number of subjects in each stage is very different. The results can be very misleading hence such a model is not considered acceptable. […] this model assumes that the formulation effect is truly different in each stage. If such an assumption were true there is no single formulation effect that can be applied to the general population, and the estimate from the study has no real meaning. Replace formulation*stage by formulation*group and you get the idea.But all of this is about a group-by-treatment interaction. Sometimes in conventional 2×2×2 crossovers we see a significant sequence (better unequal carry-over) effect. Since it can only be avoided by design (sufficiently long washout) any test for it is futile. Check for eventual residual concentrations in higher period(s) and exclude subjects with pre-dose concentrations >5% of their Cmax-values. In analogy the same can be said about the group-by-sequence interaction. Occasionally you will see a significant result. Ignore it. Only if you prefer braces plus suspenders: Go with the FDA’s model II. But no pre-test and no interaction! The loss in power as compared to the pooled analysis is negligible. Degrees of freedom in a 2×2×2 design: Model III: n1 + n2 – 2 Model III: 3(n1 + n2) – 4 What you never ever should do: Evaluate groups separately. Power will be terrible. Even if you are extremely lucky and one of them passes, what will you do? Submit only this one? Any agency will ask for the others as well. Guess the outcome. If you are thinking about the FDA’s group models I–III: There were very few deficiency letters (all with the same text) issued to US and Canadian companies. On the other hand, many European CROs have such letters collecting dust in archives. Politics? Even if you follow this track, model III (pooling) is justified since the conditions are practically always fulfilled. The sequential procedure (test for a group-by-treatment interaction at the 10% level) as any pre-test might inflate the type I error. I even have a statement from the EMA that such a procedure is not acceptable. Group-by-sequence interaction? Forget it.
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2019-08-08 12:06 (1716 d 11:18 ago) @ Helmut Posting: # 20486 Views: 7,661 |
|
Hi Hötzi, ❝ The precise model to be used for the analysis should be pre-specified in the protocol. The statistical analysis should take into account sources of variation that can be reasonably assumed to have an effect on the response variable. ❝ Is it reasonable to assume that groups or sequences have an effect on the PK response? Heck, no way! Same inclusion-/exclusion-criteria and hence, similar demographics, same clinical site, same bioanalytical method. Hundreds (thousands?) of studies performed in multiple groups and accepted by European agencies based on pooled data. But on the other hand: Why then then include e.g. period? Let us for a moment disregard the actual wording. I don't think this is about assuming that some factor has or hasn't an effect (and not about significance in the statistical sense either). As I see it I want to construct a CI which is as wide as my real uncertainty dictates. I start out with a whole bunch of ugly variation and in the fashion of Michelangelo working on his crude blocks of marble I chip parts and bits away from my bulk of variation by applying my model. What I have left of my variation is a chunk of pristine, genuine, holy, magnificent, beautiful variation. What a sight to behold , and whose origin my experimental setup cannot account for, which I therefore use for my CI. My confidence interval now is as wide as just exactly that uncertainty merits. For a crossover this is of little practical importance since subjects are in groups. For a parallel trial I think I want group in the model. If regulators don't like this, they can ask me to take it away. I happily do so without protesting. I am a sheep at that point. But not until then. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2019-08-08 12:31 (1716 d 10:54 ago) @ ElMaestro Posting: # 20487 Views: 7,694 |
|
Ahoy, my Capt’n, ❝ ❝ The precise model [] ❝ ❝ But on the other hand: Why then then include e.g. period? Cause otherwise eventual period effects would not mean out. Given, sometimes one has to assume lacking period effects and everybody is happy with that. If an originator explores whether the drug follows linear PK, we have a paired design (SD → saturation → steady state) and compare AUC0–τ with AUC0–∞. A crossover would be a logistic nightmare. ❝ Let us for a moment disregard the actual wording. [ Exactly. ❝ […] For a parallel trial I think I want group in the model. If regulators don't like this, they can ask me to take it away. I happily do so without protesting. I am a sheep at that point. But not until then. Agree again. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Astea ★★ Russia, 2019-12-21 15:12 (1581 d 07:13 ago) @ Helmut Posting: # 21011 Views: 6,947 |
|
Dear Friends! Recently I've known that Belorussian experts referring to EAEC rules require to recalculate the results of large studies using FDA's Model II, that is: Group, Sequence, Treatment, Period(Group), Group×Sequence as fixed and Subject(Group×Sequence) as Random though it was not stated in protocol beforehand. I don't think it is a good idea. Moreover it could be contagious! What do you think about it? Critical points: as standard model requires Sequence, Treatment, Period and Subject(Sequence) as fixed terms, the results of the random effect model obviously will be different. What if they will change the overall result of BE? What if the Group×Sequence effect would be significant by chance? How to deal with excluded subjects? For example, if we have a subject with only one period the fixed-effect model would automatically neglect it while the random-effect model would use it. As I understand currently it is impossible to calculate it via R, isn't it? May be Julia will help? So only people with commercial software could deal with it. — "Being in minority, even a minority of one, did not make you mad" |
PharmCat ★ Russia, 2019-12-22 02:37 (1580 d 19:47 ago) @ Astea Posting: # 21014 Views: 6,953 |
|
Hello Astea! ❝ As I understand currently it is impossible to calculate it via R, isn't it? May be Julia will help? So only people with commercial software could deal with it. I think any mixed model without replication (repeated factor) can be fitted in R and in Julia without problems. If variance components include repeated factor (structured R matrix) only special cases available in free software. |
Helmut ★★★ Vienna, Austria, 2019-12-22 11:30 (1580 d 10:54 ago) @ Astea Posting: # 21015 Views: 6,969 |
|
Hi Nastia, ❝ Recently I've known that Belorussian experts referring to EAEC rules require to recalculate the results of large studies using FDA's Model II, that is: Group, Sequence, Treatment, Period(Group), Group×Sequence as fixed and Subject(Group×Sequence) as Random though it was not stated in protocol beforehand. Regulators are always right. ❝ I don't think it is a good idea. Moreover it could be contagious! What do you think about it? Recalculating a study is never a good idea. Entire α already spent, right? ❝ Critical points: as standard model requires Sequence, Treatment, Period and Subject(Sequence) as fixed terms, the results of the random effect model obviously will be different. Possible but only to a (very) small degree. ❝ What if they will change the overall result of BE? You mean that the fixed model passes and the mixed model fails? Cannot imagine to happen (see below). ❝ What if the Group×Sequence effect would be significant by chance? Not even the FDA asks for testing Group×Sequence in Model II. ❝ How to deal with excluded subjects? For example, if we have a subject with only one period the fixed-effect model would automatically neglect it while the random-effect model would use it. Yep. On the other hand, the residual variance in the mixed model should not be larger than the one of fixed model. I had a few cases in the past where European regulators asked for recalculation according to the new GL (the mixed model was my standard before). The CI was sometimes narrower but differed only in the second decimal place… ❝ As I understand currently it is impossible to calculate it via R, isn't it? I agree with PharmCat. Doable in R. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Astea ★★ Russia, 2019-12-22 22:23 (1580 d 00:02 ago) @ Helmut Posting: # 21018 Views: 6,976 |
|
Dear Breathtakingly Smart People! ❝ I agree with PharmCat. Doable in R. Please check my solution... Below is my R-code:
library(readxl) In order to check whether it works properly I used Data set 1 from Q&A and added group-factor: first 30 subjects - group #1, and the last - group #2. To compare the results I used the following SAS-code:
proc mixed data=SASuser.Dataset; Cause I wasn't sure that my code works well I used the step-by-step approach and compared different models (they all may be meaningless by nature but whatever): 1). Sequence Formulation Period as Fixed, Subject(Sequence) as Random 2). Sequence Formulation Period Group as Fixed, Subject(Sequence) as Random 3). Sequence Formulation Period(Group) Group as Fixed, Subject(Sequence) as Random 4). Sequence Formulation Period(Group) Group Sequence*Group as Fixed, Subject(Sequence) as Random 5). Sequence Formulation Period(Group) Group Sequence*Group as Fixed, Subject(Sequence*Group) as Random The results are as follows: SAS SAS SAS R R R And now, attention, the questions: 1). Looking the same is not proving to be the same, isn't it? 2). By the way, this is exactly an example where the design changes the decision of BE (of course if scaling was not stated in protocol). BE or not BE? 3). When I try to use Subject(Sequence*Group) as Random in R, it throws an error: "couldn't evaluate grouping factor (Sequence * Group):Subject within model frame: try adding grouping factor to data frame explicitly if possible". Is it right that we can simply neglect the structure of this term as we can do in standard models with Subject instead of Subject(Sequence)? — "Being in minority, even a minority of one, did not make you mad" |
PharmCat ★ Russia, 2019-12-22 23:52 (1579 d 22:33 ago) (edited by PharmCat on 2019-12-23 17:04) @ Astea Posting: # 21019 Views: 6,996 |
|
Hi! ❝ Is it right that we can simply neglect the structure of this term as we can do in standard models with Subject instead of Subject(Sequence)? Edited. You can use just "Subject" in random statement in SAS. I think if you try to do this, you should get the same results. Of course, if your subject numbers not repeated. Data: DATA data; Code 1 PROC MIXED data=data; Fixed effect without denoted nested factors: problem with results, but coefficient calculated correctly. Code 2
PROC MIXED data=data; Fixed effect with nested factor: no problems, but Subjects in Dimensions = 1. Look at horrible V matrix. In this case it does not matter, but it not corresponds with reality. Code 3 PROC MIXED data=data; Good fit. Subjects in Dimensions = 8. We have individual V matrices and ready for structuring. Code 4
PROC MIXED data=data; Nested random part - no effect. I can't find any difference between subject and subject(room). As I understand nested factors affect on fixed part. I don't know how nested statment can affect on random part, but I think - in this case there is no effect. |
mittyri ★★ Russia, 2019-12-23 15:30 (1579 d 06:55 ago) @ Astea Posting: # 21020 Views: 6,840 |
|
Dear Nastya, whenever possible use simple solutions, at least till the moment you must dive into mixed models galaxy And more: interaction terms are not the same in SAS and R take a look at method.B implementation inside replicateBE package I just modified it a bit library(replicateBE) — Kind regards, Mittyri |
Astea ★★ Russia, 2019-12-23 18:26 (1579 d 03:59 ago) @ mittyri Posting: # 21022 Views: 6,806 |
|
Dear mittyri! Thanks a lot for your elegant solution! Didn't you think that replicateBE was the first thing that I've thought of, but I couldn't dig up the code for get.data in order to modify it for my needs. The question that I could not solve was extremely stupid: namely, what is the difference between Dataset<-rds01 and Dataset<-read_excel("rds01.xlsx", sheet = 1)? The initial dataset is the same, but the results of the algo are different. So before using we should somehow prepare the data, but how to do it? "I faced the wall...and did it my way" — "Being in minority, even a minority of one, did not make you mad" |
mittyri ★★ Russia, 2019-12-24 11:53 (1578 d 10:32 ago) @ Astea Posting: # 21023 Views: 6,760 |
|
Dear Nastya, ❝ what is the difference between Dataset<-rds01 and Dataset<-read_excel("rds01.xlsx", sheet = 1)? check Dataset1<-read_excel("rds01.xlsx", sheet = 1) — Kind regards, Mittyri |
Helmut ★★★ Vienna, Austria, 2019-12-24 12:03 (1578 d 10:22 ago) @ Astea Posting: # 21024 Views: 6,802 |
|
Hi Nastia, ❝ […] replicateBE was the first thing that I've thought of, but I couldn't dig up the code for get.data in order to modify it for my needs. I would not try to use this function. It calls others which are not exported. Not worth the efforts to modify it. ❝ […] what is the difference between Dataset<-rds01 and Dataset<-read_excel("rds01.xlsx", sheet = 1)? The initial dataset is the same, but the results of the algo are different. So before using we should somehow prepare the data, but how to do it? Duno how you accessed the datasets. Try this one:
Edit: Hey Mittyri, you were faster! — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Astea ★★ Russia, 2019-12-24 20:18 (1578 d 02:07 ago) @ Helmut Posting: # 21029 Views: 6,748 |
|
Dear Friends! Thanks a lot for your inestimable help! I finally found. The key point was just adding the string: Dataset$period<-factor(Dataset$period). Now all works the same. I still struggle to understand what is the difference between lmer and lme in this case... — "Being in minority, even a minority of one, did not make you mad" |
Helmut ★★★ Vienna, Austria, 2019-12-25 20:12 (1577 d 02:13 ago) @ Astea Posting: # 21030 Views: 6,829 |
|
Hi Nastia, ❝ I still struggle to understand what is the difference between lmer and lme in this case... The syntax of the formula for random effects is different.
lme() you have to use na.action = na.omit because the default na.action = na.fail would stop and throw an error.The method to extract the PE and calculate the CI is also different. Hint: Lines 53–56 and 80–84 of the source-code of method.B.R .— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Beholder ★ Russia, 2019-12-27 14:44 (1575 d 07:41 ago) @ Astea Posting: # 21031 Views: 6,552 |
|
Dear Astea! ❝ large studies Saying "large studies" do you mean the multicentral BE studies or large amount of subjects, e.g. 70 subjects in one clinic divided into two groups? — Best regards Beholder |
Astea ★★ Russia, 2019-12-29 22:59 (1572 d 23:26 ago) @ Beholder Posting: # 21035 Views: 6,482 |
|
Dear Beholder! ❝ Saying "large studies" do you mean the multicentral BE studies or large amount of subjects, e.g. 70 subjects in one clinic divided into two groups? I'm sorry for the wrong terminology used. I meant studies in which volunteers were divided by two groups for the capacity reasons. (Formally this can be true even for studies with 12 subjects if the clinical site is tiny.) Multicentral studies is the other case. |
Helmut ★★★ Vienna, Austria, 2019-12-30 13:32 (1572 d 08:53 ago) @ Astea Posting: # 21036 Views: 6,643 |
|
Hi Nastia, ❝ I meant studies in which volunteers were divided by two groups for the capacity reasons. (Formally this can be true even for studies with 12 subjects if the clinical site is tiny.) Might also happen in mid-range CROs with drugs which require continuous cardiac monitoring. 12–16 beds is not unusual. Nevertheless, I don’t understand why Belorussian experts ask for a mixed model because the guideline recommends ANOVA: 88. Сравнение исследуемых фармакокинетических параметров проводят с помощью дисперсионного анализа (ANOVA). Fixed effects are specifically recommended (taking into account effects which can affect the response):89. Статистический анализ должен принимать во внимание источники вариабельности, способные повлиять на изучаемую переменную. В такой модели дисперсионного анализа принято использовать такие факторы, как последовательность, субъект последовательности, период и лекарственный препарат. В отношении всех этих факторов следует использовать фиксированные, а не случайные эффекты. We discussed that ad nauseam. At least in the EU no regulatory statistician assumes that groups have an impact and expects a group-model. Pooled analysis rulez (see above). However, why all that fuzz? Let’s simulate 100,000 studies:
But you are right that in rare borderline cases a study passing with the pooled analysis might fail with the group model due to the lower degrees of freedom. \(df\text{(pooled model)}=N-2\) \(df\text{(group model)}=\sum_{i=1}^{i=groups}n_i-(groups-1)-2\) Of course, the larger the sample size and the smaller the number of groups the impact will be decrease.
❝ Multicentral studies is the other case. Correct. I would always recommend to include site-terms in the model. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |