cakhatri ★ India, 2019-01-10 08:20 (1904 d 11:16 ago) Posting: # 19764 Views: 3,887 |
|
Dear All, Happy New year I wish to have inputs on the following issue 1. We are currently working on a Drug-Drug Interaction study where Digoxin is the substrate & X is the Perpetrator. 2.During method development we observed around 12-16% interference from Digoxin deuterated (D3) IS at the RT of Digoxin drug. There seems to be no source for D5 or D7 or D11. 3. Inspite of this interference, all the QCs (100%) (Spiked with only Digoxin & Digoxin + X) are meeting the acceptance criteria. 4. Two different sources of Deuterated IS were tested and results are similar. 5. The LOQ that we have selected is 5% of reported Cmax and there is no possibility to increase the same considering guideline recommendations. 6. Digoxin forms an adduct with Ammonia. 7. The MRM pairs are Drug (798.4 / 651.4) & Digoxin D3 (801.4 / 654.4) 8. Mol.wt for Digoxin is 780.94. My question 1. Would it be advisable to proceed with the validation and will this have impact during sample analysis since all the samples will be spiked with IS. 2. Is there an alternate approach to this scenario ? Kind Regards Chirag |
VKB2207 ☆ Portugal, 2019-01-10 21:48 (1903 d 21:48 ago) @ cakhatri Posting: # 19767 Views: 3,395 |
|
Hi, Suggest you not to go with MV in this scenario as IS is contributing and you have no exact idea how much it is contributing. Hence measured concentrations would be inaccurate. You can probably evaluate the % contribution from IS at RT of drug and then apply this factor in CC, QC nominal concentrations. Hope it helps. Regards Edit: Full quote removed. Please delete everything from the text of the original poster which is not necessary in understanding your answer; see also this post! [Ohlbe] |
Ohlbe ★★★ France, 2019-01-10 23:36 (1903 d 20:00 ago) @ VKB2207 Posting: # 19769 Views: 3,327 |
|
Dear VKB2207, ❝ You can probably evaluate the % contribution from IS at RT of drug Sure. It's called the intercept ❝ and then apply this factor in CC, QC nominal concentrations. Certainly not. It is taken into consideration when calculating the calibration curve parameters, unless you are forcing the intercept through 0 (which is not to be done in bioanalysis). ❝ Hence measured concentrations would be inaccurate. Not necessarily at high level of concentrations. Could indeed be problematic closer to the LLOQ. — Regards Ohlbe |
Ohlbe ★★★ France, 2019-01-11 00:41 (1903 d 18:55 ago) @ cakhatri Posting: # 19770 Views: 3,412 |
|
Dear Chirag, ❝ 2.During method development we observed around 12-16% interference from Digoxin deuterated (D3) IS at the RT of Digoxin drug. That's a rather high level of interference. I'm quite surprised, that's more than I would expect just from isotopic contribution. It does not look like it could be due to another type of adduct, figures do not really match, but I'm not expert enough. Did you check that the deuterium substitution was not done at a labile position, causing deuterium exchange ? There are several papers describing the use of D3 digoxin as IS. Do the authors report similar levels of interference ? Was their IS marked at the same position ? Did you try and modify your ion source parameters (e.g. declustering potential) ? Any influence on this % interference ? Any improvement with an increased MS resolution ? I know it will spoil the LLOQ, but it could help to understand what's going on. ❝ 3. Inspite of this interference, all the QCs (100%) (Spiked with only Digoxin & Digoxin + X) are meeting the acceptance criteria. Yes, the interference is compensated by the intercept in your calibration curve. But I would be worried regarding the ruggedness of the method and the reproducibility of the results at low levels of concentration, close to the LLOQ. ❝ My question ❝ ❝ 1. Would it be advisable to proceed with the validation and will this have impact during sample analysis since all the samples will be spiked with IS. I would first try and understand what's happening, and reduce this % interference as much as possible. ❝ 2. Is there an alternate approach to this scenario ? If the authors of published papers report the same issues and there is no possible correction, I would either: - try and find a provider who would accept to manufacture digoxin marked with more deuterium, or C13, - or try a structural analogue IS rather than stable-labelled, - or decrease the interference by decreasing the amount of IS added to the samples. Or you can try your luck, perform a full validation and see whether it passes, but there is always a risk that any Agency the results will be submitted to will not like it. If you could keep us informed, it would be appreciated. That's an interesting topic — Regards Ohlbe |
cakhatri ★ India, 2019-01-11 10:28 (1903 d 09:08 ago) @ Ohlbe Posting: # 19772 Views: 3,340 |
|
Dear VKB2207 & Ohlbe, Thanks for your comments. I shall update on the outcome / resolution Regards Chirag |
ElMaestro ★★★ Denmark, 2019-01-11 10:49 (1903 d 08:47 ago) @ cakhatri Posting: # 19773 Views: 3,373 |
|
Hello chirag, may I ask: ❝ 2.During method development we observed around 12-16% interference from Digoxin deuterated (D3) IS at the RT of Digoxin drug. There seems to be no source for D5 or D7 or D11. Does the IS increase or decrease the area of the analyte? ❝ 4. Two different sources of Deuterated IS were tested and results are similar. What is the purity expressed as DigD3 on your CoA's? — Pass or fail! ElMaestro |