 bioanalytical validation runs [Bioanalytics]

Dear freinds
i have a question related to the bioanalytical validation runs:
According to EMA and FDA guidlines, should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?


Edit: Category changed; see also this post #1. [Helmut]
Helmut
Hero
avatar
Homepage
Vienna, Austria,
2017-12-21 14:53

@ samar sameh
Posting: # 18075
Views: 1,803
 

 QCs in all runs

Hi Samar,

» […] should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?

Counter question:

If you would not include QCs in these runs how could you know whether their results are valid?


Cheers,
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. ☼
Science Quotes
ElMaestro
Hero

Denmark,
2017-12-21 17:30

@ samar sameh
Posting: # 18079
Views: 1,847
 

 bioanalytical validation runs

Hi samar sameh,

Matrix effect and recovery can as far as I know be evaluated without the use of a calibration curve and without regulators bashing you. I am thinking there is no harm in injecting the curve, but it isn't particularly informative if the QCs pass on it in those cases. That a run is approved by way of QC's does not necessarily imply that the run is qualified for a type of experiment that does not hinge on analyte divided by internal standard.

For stabilities I think you have often a plain injection of the curve. Hence you inject QC's in such runs.

if (3) 4

Best regards,
ElMaestro

"(...) targeted cancer therapies will benefit fewer than 2 percent of the cancer patients they’re aimed at. That reality is often lost on consumers, who are being fed a steady diet of winning anecdotes about miracle cures." New York Times (ed.), June 9, 2018.
xtianbadillo
Junior

Mexico,
2018-01-18 15:31

@ samar sameh
Posting: # 18213
Views: 1,420
 

 bioanalytical validation runs

Hello Samar:

» According to EMA and FDA guidlines, should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?

QCs must be include within study samples but not in each validation run: "An analytical run consists of the blank sample (processed matrix sample without analyte and without IS) and a zero sample (processed matrix with IS), calibration standards at a minimum of 6 concentration levels, at least 3 levels of QC samples (low, medium and high) in duplicate (or at least 5% of the number of study samples, whichever is higher), and study samples to be analysed" EMA Guideline

As Helmut and ElMaestro have said there are validation paramaters that do not requires QCs (selectivity, matrix effects...)
Activity
 Thread view
Bioequivalence and Bioavailability Forum |  Admin contact
18,935 posts in 4,039 threads, 1,287 registered users;
online 7 (1 registered, 6 guests [including 5 identified bots]).

There are no such things as applied sciences,
only applications of science.    Louis Pasteur

The BIOEQUIVALENCE / BIOAVAILABILITY FORUM is hosted by
BEBAC Ing. Helmut Schütz
HTML5 RSS Feed