 bioanalytical validation runs [Bioanalytics]

Dear freinds
i have a question related to the bioanalytical validation runs:
According to EMA and FDA guidlines, should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?


Edit: Category changed; see also this post #1. [Helmut]
Helmut
Hero
Homepage
Vienna, Austria,
2017-12-21 14:53

@ samar sameh
Posting: # 18075
Views: 468
 

 QCs in all runs

Hi Samar,

» […] should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?

Counter question:

If you would not include QCs in these runs how could you know whether their results are valid?


[image]Regards,
Helmut Schütz 
[image]

The quality of responses received is directly proportional to the quality of the question asked. ☼
Science Quotes
ElMaestro
Hero

Denmark,
2017-12-21 17:30

@ samar sameh
Posting: # 18079
Views: 437
 

 bioanalytical validation runs

Hi samar sameh,

Matrix effect and recovery can as far as I know be evaluated without the use of a calibration curve and without regulators bashing you. I am thinking there is no harm in injecting the curve, but it isn't particularly informative if the QCs pass on it in those cases. That a run is approved by way of QC's does not necessarily imply that the run is qualified for a type of experiment that does not hinge on analyte divided by internal standard.

For stabilities I think you have often a plain injection of the curve. Hence you inject QC's in such runs.

I could be wrong, but…


Best regards,
ElMaestro

- Bootstrapping for dissolution data is a relatively new hobby of mine.
xtianbadillo
Junior

Mexico,
2018-01-18 15:31

@ samar sameh
Posting: # 18213
Views: 72
 

 bioanalytical validation runs

Hello Samar:

» According to EMA and FDA guidlines, should we include quality control samples in each validation run (matrix effect, Reccovery, stability,,, )?

QCs must be include within study samples but not in each validation run: "An analytical run consists of the blank sample (processed matrix sample without analyte and without IS) and a zero sample (processed matrix with IS), calibration standards at a minimum of 6 concentration levels, at least 3 levels of QC samples (low, medium and high) in duplicate (or at least 5% of the number of study samples, whichever is higher), and study samples to be analysed" EMA Guideline

As Helmut and ElMaestro have said there are validation paramaters that do not requires QCs (selectivity, matrix effects...)
Back to the forum Activity
 Thread view
Bioequivalence and Bioavailability Forum | Admin contact
17,743 Posts in 3,791 Threads, 1,106 registered users;
33 users online (0 registered, 33 guests).

The analysis of variance is not a mathematical theorem,
but rather a convenient method of arranging the arithmetic.    R.A. Fisher

The BIOEQUIVALENCE / BIOAVAILABILITY FORUM is hosted by
BEBAC Ing. Helmut Schütz
XHTML/CSS RSS Feed