 urine [Bioanalytics]

Dear all,
I would like to know the conditions of storage of urine samplings collected in a bioequivalence study.
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-08-22 14:29
(6530 d 14:18 ago)

@ ioanam
Posting: # 1010
Views: 3,775
 

 storage of urine samples

Dear ioanam!

❝ … conditions of storage of urine samplings


Ones allowing to obtain valid results for long-term stability (i.e., at least from collection of the first clinical sample to closing of the analytical database)…

Whenever possible you should use PP vials instead of glassware.
You should check in analytical method development whether the compound(s) of interest (parent/meta­bolite) bind to the surface of vials (e.g., some tricyclics stick to PP).
Urine samples freeze much faster than plasma / serum / whole blood does; you have to expect many more brocken glass vials of urine samples. If you have to use glass vials, you should allow for more than usual back-ups (and keep them in a plastic box instead of cardboard).

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Ohlbe
★★★

France,
2007-08-22 14:31
(6530 d 14:16 ago)

@ ioanam
Posting: # 1011
Views: 3,617
 

 urine

Dear ioanam,

Just the same way you store your plasma samples !

There is no guideline for this. Common sense says you should follow the storage conditions you have studied during bioanalytical method validation and for which you have demonstrated that your analyte is stable: -20°C, -30°C, -70°C... If you have no validation data yet have a look at available litterature. You have to decide on a case-by-case basis, depending on you analyte and available information.

What may raise problems is also the storage conditions during the collection phase itself. Even though this will only last for a few hours, make sure your analyte is stable under the conditions used (usually +4°C) during the period of collection.

Regards
Ohlbe
UA Flag
Activity
 Admin contact
23,428 posts in 4,929 threads, 1,696 registered users;
59 visitors (0 registered, 59 guests [including 10 identified bots]).
Forum time: 04:47 CEST (Europe/Vienna)

If I’d observed all the rules,
I’d never have got anywhere.    Marilyn Monroe

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5