NK
☆    

India,
2023-06-27 05:57
(737 d 15:29 ago)

Posting: # 23616
Views: 3,466
 

 Outlier [Study As­sess­ment]

Dear All,

Please give your valuable inputs.

We have performed 2x2 fasting BE pivotal study for USFDA after seeing pilot study results.

We've noted following from pivotal study data,
  • 90% CI for Cmax and AUC0-t was within 80.00-125.00% and for AUC0-inf upper limit was not within the BE limit, this is due to one subject (T/R was >3000%; T½ is >1000 hrs; AUC extrapolation was 99% for this subject). After reaching Tmax at 2.00 hr the concentration declined and reached BQL in 12 hours and after that few measurable concentrations was observed in 14 hr, 16 hr & 18 hr in elimination phase.
  • Similar case was NOT observed in the pilot study.
How to handle this subject’s data? Outlier? Can the regulatory accept the based on Cmax and AUC0-t results?

Can we select the time point manually for Kel calculation instead of best fit method by the tool?

Thanks & Regards
NK


Edit: Category changed; see also this post #1[Helmut]
dshah
★★  

India,
2023-06-27 10:42
(737 d 10:44 ago)

@ NK
Posting: # 23617
Views: 2,838
 

 Outlier

Dear NK!

❝ How to handle this subject’s data? Outlier? Can the regulatory accept the based on Cmax and AUC0-t results?

❝ Can we select the time point manually for Kel calculation instead of best fit method by the tool?


The CRO have SOP for handling such observation. Kindly follow the SOP and treat accordingly.


The SOP I am talking of CRO are mainly w.r.t. repeat analysis.

Kindly ask for Investigation for failed analytical runs or abnormal/unexpected results or Anomalous Value repeats.

1. An investigation must be conducted to determine the cause whenever analytical run failure or abnormal / unexpected results are observed.

b. A detailed investigation is required in case, low or below LLOQ concentration obtained in an entire period, re-analysis of a particular set of samples, poor chromatography obtained in a particular period and others, but not limited to above list.


Regards,
Divyen


Edit: Second part merged with a later (now deleted post). You can edit your original post for 24 h. [Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2023-06-27 11:56
(737 d 09:30 ago)

@ NK
Posting: # 23619
Views: 2,710
 

 Visual inspection of the fit!

Hi NK,

❝ […] T½ is >1000 hrs […] After reaching Tmax at 2.00 hr the Concentration declined and reached BQL in 12 hours and after that few measurable concentrations was observed in 14 hr, 16 hr & 18 hr in elimination phase.

It doesn’t need rocket science to tell that such a t½ obtained by an automatic algorithm (e.g., \(\small{\textsf{max}\,\textsf{R}_\textsf{adj}^2\,}\)) is plain nonsense. Recall that the A&P of the bioanalytical method at the LLOQ is 20%. Furthermore, concentrations follow a lognormal distribution. Hence, it is not uncommon to face some values >LLOQ.

❝ How to handle this subject’s data? Outlier? Can the regulatory accept the based on Cmax and AUC0-t results?

When you explore ANDAs you will find many cases were the number of AUC0–∞ is less than the one of AUC0–t. Hence, it is acceptable in principle. In the recent ICH M13A draft guideline AUC0–∞ is not a primary parameter. It serves only to assess the validity of the study. Once the guideline will be final, the FDA will update theirs.

❝ Can we select the time point manually for Kel calculation instead of best fit method by the tool?

Of course, if specified in an SOP (see Divyen’ post above). Manual selection / exclusion of data is provided in any software I know. Certara states1

Using this methodology, Phoenix will almost always compute an estimate for Lambda Z. It is the user’s responsibility to evaluate the appropriateness of the estimated value.
(my emphasis)

Similarly in Lixoft’s PKanalix2 any many others.
For transparency I state the procedure also in the protocol (see this article). In 40+ years I never received a single (‼) deficiency letter in this respect.

Since you obviously didn’t have a suitable SOP in place, you can only present a sensitivity analysis with the AUC0–∞ of the one subject excluded. Supportive would also be an outlier test (Lund’s, Thompson’s τ, Cook’s distance, …) and/or graphical EDA (box plots, QQ-plots, histograms, scatter plots of studentized model residuals). See also this post about the method recommended by Health Canada. Though reanalysis based on PK grounds is not acceptable according to current bioanalytical guidances, I would do it anyway – not only the suspect values but the entire profile. Discuss it, and cross fingers.

I strongly recommend to update your SOP – mandating visual inspection of fits (see references 3, 5, 8, 20 in the article linked above) and allow to exclude data in the estimation of λz. Then you don’t have to exclude the nonsensical AUC0–∞ obtained by an automatic algorithm and keep the subject with a realistic AUC0–∞ in the analysis.
Maybe over the top, but I recommend also to perform the entire NCA in a blinded manner (see this thread). Avoids the impression of cherry-picking.



  1. Certara USA, Inc. Princeton, NJ. 7/9/20. Lambda Z or Slope Estimation settings. Online.
  2. LIXOFT, Antony, France. 2022. PKanalix Documentation. Check lambda_z. Online.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,426 posts in 4,929 threads, 1,682 registered users;
40 visitors (0 registered, 40 guests [including 25 identified bots]).
Forum time: 21:27 CEST (Europe/Vienna)

You can do one of two things; just shut up,
which is something I don’t find easy,
or learn an awful lot very fast,
which is what I tried to do.    Jane Fonda

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5