Moloko
☆    

Russia,
2020-12-22 12:41
(1657 d 02:03 ago)

Posting: # 22153
Views: 4,363
 

 Significant effects [General Sta­tis­tics]

Hi all,

Could somebody help me with my doubts about significant sequence/treatment/period
effects in BE study? I got the following anova results in 2x2x4 design:

                 DF Sum Sq Mean Sq      F   P-val
sequence          1  0.558   0.558 10.269   0.002
period            3  0.566   0.189  3.472   0.021
treatment         1  0.075   0.075  1.384   0.244
sequence:subject 23  7.632   0.332  6.102  <0.001
residual         69  3.752   0.054


My СI-s are within the range of 80.000% - 125.000%, so formally BE acceptance criterion is met. But these p-values scare me... :confused:

I don't have much experience in BE and I think it is not good then period or sequence factor influences PK-parameter.

thanks!


Edit: Category changed; see also this post #1. Tabulators changed to spaces and BBcoded; see also this post #6. [Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2020-12-22 17:27
(1656 d 21:17 ago)

@ Moloko
Posting: # 22154
Views: 3,531
 

 Significant effects: Don’t worry, be happy!

Hi Moloko,

❝ Could somebody help me with my doubts about significant sequence/treatment/period

❝ effects in BE study?


Search the forum for sequence effect, treatment effect, and period effect. Lots of posts, references, etc. See also the articles about carryover and a significant treatment effect.

❝ My СI-s are within the range of 80.000% - 125.000%, so formally BE acceptance criterion is met.


Excellent. Open a bottle of champagne. :party:

❝ But these p-values scare me... :confused:


❝ I don't have much experience in BE and I think it is not good then period or sequence factor influences PK-parameter.

  • Unless the the study was extremely imbalanced (say, almost all subjects in one sequence and the rest in the other), period effects mean out, since both T and R are affected to the same extent. Nothing to worry about.
  • Sequence effects (actually unequal carry-over) cannot be “corrected” by a statistical method – only avoided by design (sufficiently long washout). It was shown in a large meta-analysis that in properly designed studies a significant effect is seen in ~ the level of the test and hence, likely is a statistical artifact (false positive). Forget it.
  • Even a significant treatment effect is not important. Only the CI-inclusion counts.
  • The fact that subjects differ is trivial. Only if you don’t see a significant effect you should start to worry (monozygotic twins or triplets in the study?) because it would violate the assumption of independence.
See also here for the sequence effect and there for the treatment effect. The period effect cancels out in a crossover study anyway.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Moloko
☆    

Russia,
2020-12-23 14:24
(1656 d 00:20 ago)

@ Helmut
Posting: # 22158
Views: 3,423
 

 Significant effects: Don’t worry, be happy!

Helmut, thank you very much for response!

It seems like I couldn't find the correct keywords when searching for posts earlier..
So now I can immerse myself in reading:-)
UA Flag
Activity
 Admin contact
23,427 posts in 4,929 threads, 1,676 registered users;
44 visitors (0 registered, 44 guests [including 11 identified bots]).
Forum time: 15:45 CEST (Europe/Vienna)

Many people tend to look at programming styles and languages like religions:
if you belong to one, you cannot belong to others.
But this analogy is another fallacy.    Niklaus Wirth

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5