Calibration Curve [Bioanalytics]

posted by Ohlbe – France, 2012-01-21 14:44 (5255 d 05:24 ago) – Posting: # 7986
Views: 5,931

Dear Auditor,

Sorry I had misunderstood your question, I thought you were asking which point to use for calculations. Regarding plotting usually HPLC or LC-MS/MS software can plot all points, whether used or not. Points excluded from the calculations will look differently (e.g. clear circles instead of black dots). I would plot all points, including unused duplicates. Don't plot means, unless you are using means for calculations.

Regards
Ohlbe

Regards
Ohlbe